Upgrading Earlier Versions - Instruments with firmware version V5.0x and V6.x cannot be upgraded to V8.04 in the field. These instruments require re-calibration after upgrading. Therefore these units must be returned to a Fluke Authorized Service Center. Contact your local service center and order a firmware upgrade and calibration according to PCN 4841. Note: For the firmware upgrade and calibration a nominal fee will be charged.
That's probably because someone update to the latest firmware. I remember in the old Fluke Forums that there was a warning:QuoteUpgrading Earlier Versions - Instruments with firmware version V5.0x and V6.x cannot be upgraded to V8.04 in the field. These instruments require re-calibration after upgrading. Therefore these units must be returned to a Fluke Authorized Service Center. Contact your local service center and order a firmware upgrade and calibration according to PCN 4841. Note: For the firmware upgrade and calibration a nominal fee will be charged.Going backwards is not possible too, because the calibration was lost already.Try this from the service manual (Fluke 192B/196B-C/199B-C Service Manual 5-18) - http://assets.fluke.com/manuals/192B____smeng0100.pdf (Attachment Link)