fluke 287 dmm
« on: January 16, 2017, 01:10:32 pm »
hi all     i have fluke 287 dmm .now not measurring Vlotages & current. showing on disply "lead connect correctly".can u helpme repairng this
 

Offline mzacharias

  • Frequent Contributor
  • **
  • Posts: 709
  • Country: us
Re: fluke 287 dmm
« Reply #1 on: January 16, 2017, 01:59:12 pm »
Has the meter been exposed to moisture or other contaminants? If the current jack sensing is triggered it will cause this symptom.
 