I received a dead 289 a while ago, saw a blown cap on the board, replaced it, didn't fix the problem, and so it went in a drawer, where it remained for years.
Yesterday, I somehow unintentionally came across the meter, and thought I'd give it another try...
I immediately noticed the batteries were getting hot quite quickly (the meter was disassembled in the drawer)... on the board, the (+ & -) pads had a 'normal' resistance between them, but once the battery holder was installed, I almost had a short (just a few ohms) on the holder's pads !
Long story short, the problem was that the replacement cap is somewhat taller than the original, and has conductive pads on both sides (all around), which shorted with one of the holder's flat pads (see pics). The battery holder has in fact 3 contacts (45° inclination) that match those on the board, but also 3 'flat' pads, that get pretty close to the board once installed. Putting some tape on those just to test, fixed the problem.
Given the layout of the board, I absolutely don't get it

What the heck are these flat pads good for ? They get damn close to components on the board, and there's no insulation whatsoever... doesn't seem like a sound design to me, but on the other hand, Fluke is typically doing things right... so am I missing something here ?
Comments welcome !