Check the board under high light, is there any problem with the paths?
Check the rotary pins.
Check RT1- RT2- Z1 and R4.
i have traced the part of the track and there are no parts that have corrosion or problems.
then the rotary PIN also looks fine, clean, there is not a lot of dust stuck to it or blocking the switch.
RT1, RT2, and R4 value is 1KΩ ±1.024KΩ to 1.026KΩ i measured with fluke 175, but for Z1 i don't understand how to measure it? if it's just a resistance, i get a value closer to 1MΩ.
===============================
Edit:
now i have removed all the components that i think are connected to the red terminal, and i found that when one is removed, namely Z1 or RT1, the multimeter can display zero in voltage menu, and display OL in the Ohm menu, but the problem is not resolved, the damage is not in RT1 or Z1 they are fine, but i have to find again the path that i marked with a red arrow, because i see there is still pseudo resistance on the path marked by the arrow maybe it should be in the resistance measurement that there should not be any value with COM, and i haven't found out what component is problematic there.