| frenky:
Some research... Fluke 287/289 are using RedBoot/eCos: --- Quote ---eCos and RedBoot based products showcase Industrial and Automotive Fluke 289 and 287 True-RMS Multimeters. Fluke's "The Most Advanced Multimeters, Ever", data logging & graphing multimeters. --- End quote --- https://lists.ecoscentric.com/ecos/examples.shtml So I have been reading alot about RedBoot and found this: --- Quote ---Accessing RedBoot After all connections have been made, start the serial terminal application and apply power to the board. Press Ctrl+C several times on the serial terminal until you see the RedBoot> prompt. If the kernel begins to boot, reset the board and try again (the standard timeout before booting the kernel is one second). On most versions of RedBoot, you will see either a '+' or a message from RedBoot before it boots the kernel, but some configurations have this disabled. --- End quote --- http://wiki.emacinc.com/wiki/Loading_Images_with_RedBoot So the reason that I could not write to console is that I was trying to do it after the kernel has already booted. So I'll try sending in Ctrl+C every half a second and then power up the Fluke 287. Hopefully I'll be able to access console this way... |
| purpose:
Incoming. |
| purpose:
9V TO PADS, IN AC V MODE, ALL READINGS RELATIVE TO NEG PAD. TP 1 0.005 2 3.11 3 4.99 4 3.11 5 3.09 6 3.09 7 0.00 8 9.00 9 2.30 10 0.00 11 0.158 12 0.00 13 3.06 14 0.00 15 2.94 16 2.909 17 3.11 18 1.82 19 2.94 20 2.94 21 2.94 22 2.94 23 2.94 28 0.00 30 0.00 31 0.00 32 3.11 33 3.11 34 0.22 Off state showed only residual and dropping V readings. Your pic answers. tp3 4.99 x1 0.00 x2 0.903 x3 2.941 x4 0.910 x5 1.820 More to follow. |
| purpose:
Tants 61 17.55 58 1.245 56 2.493 59 20.03 62 18.79 60 8.99 57 9.00 53 8.69 51 9.00 44 4.98 32 2.50 26 2.50 20 4.98 89 8.99 80 1.82 68 1.82 79 2.94 67 2.94 68 1.82 41 2.50 42 3.11 35 3.11 Peter |
| frenky:
Wow thank you. You are the best! :-+ I'm sure that many (including me of course) will find your effort realy helpful while trying to fix faulty Fluke 287/289. :-DMM |
| Navigation |
| Message Index |
| Next page |
| Previous page |