On the above wave forms the output was set to 50.00000. The manual is contradictory, it says to put range to 1000 but output to 50.0000. I tried boith and did not see a noticable difference in the waveforms.
I am not sure how to hook up the oscilloscope with the ground on + sense. Would I then hook up the input to - sense? Would I disconnect the shorting bars?
I did hook the scope with the ground on - and input on + with shorting bars present between sense and output. I am not sure I see anything meaningful. I do see a 30-50Khz ish (fluctuating) spike with the scope sweep speed set high, (don’t recall how fast exactly).
On the HP 3456A I do note that as I increase the output the noise seems to travel with the increase in output. That is, if I set the output of the 332D to 1.000000 I see noise in the uV range, if I set the output to 100.0000 I see noise in the mV range. (Nosie is amplified?)
There is also a noticeable buzzing coming from the rear of the unit (HV mother board, pre reg, or series pass) that sometimes becomes louder for a brief period. I do not notice any correlation to the increase in this buzzing and noise on the output. I am wondering if other units exhibit this buzzing sound.
My 335D hums/buzzes when feeding high(ish) currents especially when in current limit. I very rarely have it above 100V but it probably makes noise when outputting high voltages.
Don't disconnect the sense terminals from the output terminals except to regulate voltages at the input terminals of a load.
Be very aware of the scope grounding; the ground prong on the plug at the wall socket is connected to the test lead 'ground/common'. You can get into very serious trouble if you isolate the scope case/chassis (assumed to be metal). Touching it or any other exposed metal (USB socket, BNC sockets, common of any other channel, etc.) while the scope common is connected to a voltage source can be fatal. If the scope is connected to anything (computer, scope cart, terminal whatever) the scope chassis common voltage is coupled thru to that device. There are many opportunities to let the magic smoke escape or worse.
The ground prong on the 332 is connected to the outer case*/front panel BUT NOT TO THE INNER CHASSIS WHICH IS CONNECTED TO THE + OUTPUT/SENSE! The + output is the common terminal for the various control power supplies. You need to think of 'ground' and 'common' (and 'Guard') as different nodes that can sometimes have very high voltages between them. It is possible to ground (to earth) either the + output terminal or the neg output terminal or neither of them.
To do what Dr Frank asks you need to make sure that both of the 332 outputs are isolated from ground or that the + terminal is grounded. That way you don't have to float the scope.**
If you examine the mechanical attachment of the front panel to the inner chassis, you will see that the guard terminal is connected to a metal strip between them and there is a rather complex stack of metal and insulation. It's all arranged so that the front panel (and the screws attaching it) is grounded and the guard intercepts any leakage current between the inner chassis to the front panel. I think the guard is also connected to shield terminals on the transformers. For most usage, you can put a shorting bar between the guard and ground terminals on the front panel. I think the guard may be connected to the shield inside of the case but haven't checked to see if that is true.
My 335D had a shorting bar between the ground terminal and - output/sense terminals.
*The outer case is only grounded when the inner chassis is fully inserted. Otherwise it is floating. Same is true for the inner shield which may be connected to the guard terminal )or ground, I don't know which) when the chassis is fully inserted.
**Edit: You DO have to float the scope to examine anything in the pre-regulator circuit. It has direct ohmic connections to the power plug and is not isolated from the power outlet <by a transformer>. In that case power distribution wires labeled 'line (hot)', 'neutral' and 'ground' must be considered wrt the scope connections.
Edit again to add <by a transformer>