Update #2:
My next step will be to tear down the front end of the meter to gain better access to the "sample string" so I can test and record each and every resistor in the string. Based on the units behavior, I have a feeling that once I find and replace the open resistors, this unit may be ready to go back into service (after calibration, etc).
Note: I did get some cool photos and will update this post with the pics shortly.
Don't disassemble the front. you might damage or pollute the string assembly.
To check the resistors, simply measure the resistance from beginning to end of the string, and dial through the different resistor settings, either in 1-2-2-2-2-2 or 1-2-4-4 sequence for each decade, depending on the revision of your instrument, see manuals. You need to pull A4 & A5 first.
Pictures and further hints are already available, search for '.. 332B in the slaughterhouse..'.
I doubt that the string is the problem, because if one resistor is defect , other settings will give a precise reading, anyhow.
If you have the old mechanical chopper amplifier, there's probably your problem, also the HV generation card may also fail.
Frank