On the lower Ohms range it shows about 12 Ohms too high.
The error decreases slowly, up to about 4700 Ohms where the error is gone.
Continuing up towards 10K it starts showing a too low resistance.
The measurement current is 740 uA.
On serveral other 45's the current is 800 uA, and the resistance measurements are spot on.
Thinking current may be lost/shunted on the way from the current generator in the Fluke ASIC,
I have measured the voltages across resistors on the way to the front panel.
All mesurements are consistent, and it seems the ASIC outputs 740 uA.
Perhaps the current is not the issue here.
I did however find it peculiar that all others (I have another five instruments) use 800 uA.
I have tried a front panel calibration,
which slightly improved the situation, but the issue remains.
In all other respects the instrument measures correctly, Volts, Amps, DC, AC etc.
Any ideas, anyone?