Author Topic: Fluke 73 Series III service manual  (Read 1626 times)
Offline LateyMateyTopic starter
« on: August 16, 2020, 07:34:19 pm »
Hi Guys
I have a Fluke 73 Series 3 which does not read DCV correctly (actually, it's an eBay purchase which I've not received yet). Just wondered if anyone has the correct service manual I may obtain before it arrives. The series 2 one seems readily available, but not 3. Is this because its still repaired by Fluke perhaps?
I work as a Calibration Engineer here in the UK, having previously been a production and service engineer so will probably be OK without, but they do make life so much easier, don't they?
Many thanks in advance.
Leighton
Offline shakalnokturn
Re: Fluke 73 Series III service manual
« Reply #1 on: August 16, 2020, 09:45:43 pm »
I think the 23 II service manual is very close.
Offline retiredcaps
« Reply #2 on: August 17, 2020, 03:45:01 am »
When you get the meter, come back and post the problem.  We can help.
A non Fluke resource that you might find helpful.
http://mrmodemhead.com/
« Reply #3 on: August 20, 2020, 07:17:57 am »
Thanks for your replies guys.
Just as an update. The meter has arrived, its actually a Fluke 75 series III. When it came, I checked it over with the supplied fused probes and nothing, but the meter powered on OK. Swapped the leads from another meter and it seemed to work fine so I took it into work yesterday and gave it a full calibration. Everything was inside specification apart from the 32 Meg range so a bargain meter for the bench.
Thanks anyway  8)