EEVblog Electronics Community Forum
Electronics => Repair => Topic started by: hitecelec on November 09, 2019, 10:01:59 am
-
hi
looking for the commercial part number for fluke LCD Display and service manual
picture attached
thanks in advance
Anand
-
Directly from Fluke's calibration manual - LCD MODULE, 240X200 GRAPH TRANSMISSIVE - 602430.
https://dam-assets.fluke.com/s3fs-public/74x_____cmeng0200.pdf (https://dam-assets.fluke.com/s3fs-public/74x_____cmeng0200.pdf)
https://www.ebay.com/itm/for-multimeter-F-744-FLUKE744-FLUKE-744-LCD-screen/123686026086?hash=item1ccc42fb66:g:LLwAAOSw0dFchhuW (https://www.ebay.com/itm/for-multimeter-F-744-FLUKE744-FLUKE-744-LCD-screen/123686026086?hash=item1ccc42fb66:g:LLwAAOSw0dFchhuW)
-
The polarizer film went bad, its a characteristic of some vintagem LCDs from 90s, just replace both (easily found online).
Carefully disassemble the display, peel of the old polarizers
Clean the glue residue, try rubbing alcohol, then acetone and last gasoline
Make sure the screen is totally residue free, no oils or dirt
Play with the film to find the best contrast angle. Use a scalpel to cut the film, not a scissor (they crack the film)
That was my THS720P before:
(https://www.eevblog.com/forum/repair/tektronix-ths720p-bad-screen-lcd-lm32p10/?action=dlattach;attach=168119;image)
(https://www.eevblog.com/forum/repair/tektronix-ths720p-bad-screen-lcd-lm32p10/?action=dlattach;attach=168121;image)
And after replacing the polarizer. You can see I missed cleaning the backlight diffuser.
(https://www.eevblog.com/forum/testgear/tek-ths720a-portable-scope-teardowndiscussion/?action=dlattach;attach=852746;image)
-
hi
thanks for all the hints. Will try out and let you know