POR = power on reset, so if I switch it on first time. On TP6 I see Vin, after U7 is see a charge/discharge curve depending on Vin but after comp Ü9 and input pin38 of uC I only see the short spike from 3rd picture. So, "somehow" the changing charge/discharge curve is not affecting the pulse/pause at pin38 - I assume that it should so? I was thinking about testing the uC beside by using a duty cycle control of a 2Hz PWM and a optokoppler instead of U9... (just to see IF it WOULD work, if the ADC is up again).
EDIT: Yes, it shows "1 . " over all modes and ranges, so at kleast nothing works but that seems OK without the "bootleneck" of the ADC running?
Klaus.