Because of a hint in the TEA thread i got a 5 1/2 digits (119999) fluke 8375, which of course only showed zeros in the display.
For the 8375 i did not find any manual, but the 8300 manual works fine. The 8375 seems to be a 8300 with 10 ohm, 100mV and true rms voltage options (numbers are ranges, resolutions are eg. 1 µV, 100 µOhm) and a better reference (0.002 % vs 0.005 %).
The reason for the zeros display was a short between two strip lines on the main pcb which where parallel through the whole pcb about 40 cm / 16 inch. Of course there was at first sight nothing visible. So i used a 4-wire milliohm capable meter, connected the 2 source wires each on oposite sides of the two strip lines and measured the resistance between 2 parallel points with the 2 sense lines. Think of a H. In the middle is the lowest resistance. So i found the position with the lowest resistance.
And on further inspection i found a needle forest of ultra thin tin whiskers of about 2mm length (1/10 inch). After removing them with a rubber the instrument went to life.
Reason for tin whiskers are many. In this case it seems copper lines with tin plating and no other coating between.