Fluke 8506A Repair
« on: August 11, 2016, 07:12:21 am »
The repair section is below, this first part covers some of the features of the meter.

About a year ago I picked up a Fluke 8506A on a whim off of Ebay. A 6.5 digit meter selling for less than a hundred dollars, I though why not. I already had A Keithley 2701 and needed a second bench top meter to simplify things when building circuits and trouble shooting but didn't want to spend an arm and a leg to get a newer meter.

Other than knowing that it was a "6.5 digit meter," that presumably did DC and AC volts (and maybe ohms or DC amps if your lucky enough to have one of them installed) I honestly didn't know much of anything about these meters when I bought it. After extensive use and comparison with my Keithley meter,  I came to odly like this quirky multimeter. While rather large and slow behind the wheel, Fluke's 8506A DMMs are an interesting mix of 1980s metrology grade and general purpose features.

The DC voltage measurement accepts 100% over-range on the 100mV, 1V and 10V ranges, while the 100V and 1kV ranges are limited to a maximum input of 128V and 1.2kV, respectively. This feature gives a much appreciated resolution increase for readings below 20V, when compared to non over ranging DMMs. Further resolution can be had on the 10V range when averaging is enabled, which extends the 10V range to 7.5 digits. This is one of the reasons I decided to keep the 8506A for my work bench. Another neat feature is that you can independently zero correct each range of the DC volts and ohms functions and even save these corrections if you enable the cal switch on the back of the unit. Sounds boring, but it makes measurements that require changing between ranges more accurate.

Unfortunately, the DC amps, AC Volts and Ohms measurements are all 5.5 digits. However, with averaging enabled they extend to 6.5 digits but this slows the measurement speed down considerably, and this is especially true for hi-accuracy AC volts measurements which are already slow to begin with.

The AC voltage measurement uses thermal converting methods (presumably why it's so slow) and can measure up to 8:1 crest factor signals with full accuracy. 3rd dB bandwidth is excellent for a DMM with 3MHz on the 100mV range and 10MHz for the 300mV, 1V, 3V, and 10V ranges. Which is similar to the HP 3400A analog thermal converting voltmeter I also have on my bench that gives 3 digit readings but is capable of handling much larger crest factors.

Another neat feature is that you can do AC+DC TRMS volts measurements by pressing DC volts and one of the AC volts modes at the same time. My Keithley is only able to do DC or AC volts measurements, thus you would have to manually calculate the AC+DC TRMS value by hand when using it.

If installed, the DC amps module is quite handy with 100uA, 1mA, 10mA, 100mA and 1A ranges. It may not measure AC current or as high of DC current as most modern meters, but with small low power circuits it's more than adequate. I've frequently use the 8506A to measure currents smaller than 100mA, as my Keithley can't go that low.

The Repair:

A few months back I acquired three "dead" Fluke 8506A DMMs at my university's surplus sale and was curious if I could repair them and maybe pass them on to people who would actually use them, rather than let them become landfill fodder. With procrastination in full force, it wasn't until this afternoon that I finally had gotten around to giving them a rigorous tear down.

All three units turned on but gave different errors and symptoms. One read error "E" which means that an unallowed AC volts module is installed (i.e. average/RMS modules from the 8505A, the 8506A's little brother). It turns out the control module mistakenly identified the meter as a 8505A instead of a 8506A. Removing the AC volts module corrected this but then the meter latched on error 5, which corresponds to the  recirculating remainder ADC module. Swapping it out with two other units returned the same error, so it's likely something else.

The second unit turned on with no error codes but gave spurious readings on the 1000V and 100V ranges (fluctuating around 600Vdc open circuit and 100V short circuit). The third meter latched on error code 6, which is a display overflow. Turns out meter two and three suffered from the same problem but with varying degrees of severity. Looking at the manual, you would believe that the DC signal conditioner, recirculating remainder DAC, or possibly the active filter were faulty. However, after randomly swapping U22 and U20 (both EEPROMs, but I didn't know that at the time) from meter one into meter three, I was able to get meter three to respond normally (no more overflow or spurious readings). At this point, I used meter three, to systematically test the modules and power supplies from all three units for faults.

Bizarrely enough, not a single module or power supply was determined to be faulty, each one measured and operated as intended. That is, except for the control modules. The control module for meter one was stuck thinking it was an 8505A. *Note that every module is interchangeable/shared between the 8506A and 8505A except for the AC volt modules.* Since it would be time consuming to figure out why this module thinks he's a dude who's playing a dude who thinks he's another dude (Tropic Thunder reference), I decided to work on why module two wasn't working.

Turns out module two is for the CT variant of the 8506A, a military oriented model. Further, it seems this variant is able to measure up to 1000V AC instead of the 500V AC of the non CT models and extends the 100mV AC reading to 6.5 digits when averaging is enabled. This is accompanied by a few changes to the input board, the rear input switch and rear input connector are deleted and in their place, two trimmer caps and associated compensation network were added to the input board that the non CT models don't have. Thus the chassis is slightly modified to accommodate adjusting these new trimmer caps and also means that the controller module for the CT is not compatible with the non CT models. Consequently, I wasn't able to place control module two into meter three for testing (it just latched on an error), instead I had to repopulate meter two.

On a whim I decided to exchange U20 and U22 with control module three (that I originally took from control module one) and like magic meter two was working A-okay!. Upon further inspection, only U20 needed to be exchanged for meter two to work. And with further testing I was able to repeatably recreate the original errors by swapping the faulty U20 and known good U20 into each respective meter with all their original modules installed. Thus it was a control module failure on meters two and three and not a DC signal conditioner problem that I had initially thought.

Now the damnedest thing about all of this is that U20 is a Xicor X2804AP-45 4kbit (512x8) EEPROM, why this would cause the 1000V and 100V ranges to have overflow/spurious readings but not the 100mV-10V ranges is beyond me. Truly some black magic that escapes my understanding. I would think that the meter would latch on an error for the control module, but instead it still turns on and attempts to work with the faulty EEPROM.

At any rate, I hope that anyone who is working on a 8506A or 8505A and observes spurious readings that would imply a bad DC signal conditioner comes across this and tries swapping control modules (if a second unit is available) or swap U20 on the control module before looking elsewhere. While I don't know how often this problem is statistically likely to happen, having got three meters each with faulty control modules seems to imply that the control module is a common weak point, and is likely a good place to look for faults when trouble shooting.

Since I feel reasonably accomplished at the moment, control module one will have to wait until I feel up to burning another day to discover why it has mistaken identity syndrome.

When I get a chance I'll upload pictures of the units and some of the more interesting boards. The DC signal conditioners and DAC's use custom made hermetically sealed resistors and later units like the CT model I have (chip date codes point to 1993 MFD), use laser etched resistors.
« Last Edit: September 01, 2016, 06:25:03 am by garrettm »
 
The following users thanked this post: doktor pyta, evb149, SilverSolder, Barbouri, Storm, mdubinko

Offline Vgkid

  • Super Contributor
  • ***
  • Posts: 2727
  • Country: us
Re: Fluke 8506A Repair
« Reply #1 on: August 11, 2016, 02:30:13 pm »
I always liked those 850X series multimeters. The perplexing errors always shyed me away from them, eventually I got a 3456A as my first* real 6.5 digit dmm.
The foreverly broken 7065 was my first. *
If you own any North Hills Electronics gear, message me. L&N Fan
 
The following users thanked this post: kado

Offline whitevamp

  • Regular Contributor
  • *
  • Posts: 158
  • Country: us
Re: Fluke 8506A Repair
« Reply #2 on: August 11, 2016, 03:01:57 pm »
I have a 8502A and it had quit working and wile looking for the answer i found out that the dc card was out of a 8505/6. so at least the some of the cards are back wards compatible.
still need to fix my 8502A thought, but that one is beond me. |O
 

Offline garrettmTopic starter

  • Frequent Contributor
  • **
  • Posts: 279
  • Country: us
Re: Fluke 8506A Repair
« Reply #3 on: August 11, 2016, 09:29:42 pm »
I have a 8502A and it had quit working and wile looking for the answer i found out that the dc card was out of a 8505/6. so at least the some of the cards are back wards compatible.
still need to fix my 8502A thought, but that one is beond me. |O

I would set up an email alert with ebay to notify you when any 8502As are listed and then wait for one to come around your price range, likely <$100 item+shipping. Then do a systematic check of the modules from the two units. After that process you will likely have identified the problem module. And if you are lucky, you just use a module out of the new unit, if it doesn't already fully work outright. Then just resale the leftover unit for parts for someone else to use or sell off the known working modules out of it and scrap or junk box the rest. (Sometimes the chassis is destroyed and isn't worth reselling, but the insides still work and are sellable.)

If most of the modules are interchangeable with the 850xAs, I have a bunch on hand as I plan to scrap one of the 8506As due to so much damage to the chassis (it's been beat the hell up). So if you know which module you need to replace I could sell it to you on the cheap like $10 + whatever shipping costs.
 

Offline garrettmTopic starter

  • Frequent Contributor
  • **
  • Posts: 279
  • Country: us
Re: Fluke 8506A Repair
« Reply #4 on: August 11, 2016, 09:51:55 pm »
Here're some pictures of the control modules:



Module three has the AMD processor in the middle and is the oldest dated at '83. The top is module one and is dated at '87. And finally, module two is the bottom one dated at '93 and is the CT variant which uses v1.0.1 for the firmware, the others use v6.0.6 for 8506A use and v5.0.5 for 8505A use. The newest 8506A firmware that I've seen is v6.0.7 on my original 8506A that dates to '93 as well.



Close up of the non CT control boards, note that U22 was swapped between module one and three, so the date code on that chip doesn't correspond to the board that it is populated in.



Here's a close up of the two faulty EEPROMs. The cerdip from '83 has faster timings (350ns) while the pdip parts from '86 and '93 are slightly slower (450ns). The data sheet can be found here.



Heres a close up the working EEPROM



Control module three installed back into the meter with the working EEPROM from module one.



Meter three now no longer shows spurious readings with the working EEPROM. I didn't think to take pictures of the display with the faulty EEPROM at the time, but I might upload one later when the replacement EEPROM arrives.
« Last Edit: August 12, 2016, 08:33:10 pm by garrettm »
 
The following users thanked this post: evb149, Storm

Offline whitevamp

  • Regular Contributor
  • *
  • Posts: 158
  • Country: us
Re: Fluke 8506A Repair
« Reply #5 on: August 12, 2016, 01:35:02 am »
I have a 8502A and it had quit working and wile looking for the answer i found out that the dc card was out of a 8505/6. so at least the some of the cards are back wards compatible.
still need to fix my 8502A thought, but that one is beond me. |O

I would set up an email alert with ebay to notify you when any 8502As are listed and then wait for one to come around your price range, likely <$100 item+shipping. Then do a systematic check of the modules from the two units. After that process you will likely have identified the problem module. And if you are lucky, you just use a module out of the new unit, if it doesn't already fully work outright. Then just resale the leftover unit for parts for someone else to use or sell off the known working modules out of it and scrap or junk box the rest. (Sometimes the chassis is destroyed and isn't worth reselling, but the insides still work and are sellable.)

If most of the modules are interchangeable with the 850xAs, I have a bunch on hand as I plan to scrap one of the 8506As due to so much damage to the chassis (it's been beat the hell up). So if you know which module you need to replace I could sell it to you on the cheap like $10 + whatever shipping costs.

yea i have been watching ebay for the 8502/5/6 but when i see one in my  $ $ range its >100, that has the correct moduels. see i have all ready ( or at lest im pretty sure. ) identify what card is bad, and that the ohms card and the dc card.
and as for the cards individualey there was one dc card for 30 and when i finely got the $$ something came up and i had to spend that on something else, and now there not one, and well the ohms card the only one i have ever seen on fleabay has been 100, which is way out of my range, right now.

If most of the modules are interchangeable with the 850xAs, I have a bunch on hand as I plan to scrap one of the 8506As due to so much damage to the chassis (it's been beat the hell up). So if you know which module you need to replace I could sell it to you on the cheap like $10 + whatever shipping costs.

and yea that would be great if you have ether one of those cards. and the power switch (on/off/)
yea if i did get one on fleBay i wouldn't care what the case looked liked as mine is like new, no scratches/dings/sticker residue, etc. as long as it had the right cards, and working.
« Last Edit: August 12, 2016, 01:58:14 pm by whitevamp »
 

Offline Vgkid

  • Super Contributor
  • ***
  • Posts: 2727
  • Country: us
Re: Fluke 8506A Repair
« Reply #6 on: August 12, 2016, 02:11:06 am »

The HP 3456A is an excellent meter with its good DC volts stability and offset compensated 4-wire ohms. The only reason I decided against buying one of these was because of the fan, mainly since I didn't want to hear a rocket ship when using the DMM, as HP isn't known for using quiet fans. They love to use those all aluminum 120V AC fans that are built like a tank, but are equally as loud as one.
the 3456a doesn't have a fan(mine doesn't*). I do know about those Al Papst fans that you mentioned. My hp 5335 had one swapped it for an equally specced much quieter Delta.
If you own any North Hills Electronics gear, message me. L&N Fan
 

Offline garrettmTopic starter

  • Frequent Contributor
  • **
  • Posts: 279
  • Country: us
Re: Fluke 8506A Repair
« Reply #7 on: August 12, 2016, 06:13:14 am »
@Vgkid, I may have been mistaken, but I thought that the HP 3456A had a fan in them. Do you happen to have a newer unit with a white painted HP logo and the revised buttons? HP's product lines extend for decades, so it could be that the older units had fans?

For example, the HP 3400A began selling in 1963 and continued up untill the '80s (possibly '90s?) when it was replaced with the 3400B that was much shorter lived. Over the decades HP made incremental upgrades to the design of the 3400A, moving from optical choppers and nuvistor inputs to an integrated circuit based chopper and fet input which was retained but all placed on a monolithic board for the 3400B. The 3456A seems to follow a similar trend, with incremental modifications and upgrades to the base design over its product life. But that's all speculation on my part, they may have never had a fan in them and I was simply mistaken.
« Last Edit: August 12, 2016, 06:22:54 am by garrettm »
 

Offline MadTux

  • Frequent Contributor
  • **
  • Posts: 793
Re: Fluke 8506A Repair
« Reply #8 on: August 12, 2016, 05:22:42 pm »
Older HP-3456A had a fan in the back but it was removed relatively early.
Easy to identify by the round plastic extension in the back that has a fine metal filter mesh in it.

There were no major hardware revisions on the analog board, only thing that got changed over the years are the digital board (exchanged the unreliable Mostek Mask-ROMs with some higher density EPROM) and the beefy protection diodes on the interface board below the analog board were removed. Early units also had funny red chip sockets for the microprocessors that were somewhat unreliable. #1 failure mode are the capacitors in the PSU however. The rest is rock solid or easy to repair.
The fan is relatively easy to remove, so it's probably no problem to simply remove it, if a fan has bad bearings and gets noisy.

I personally don't really like the newer silicone rubber keypads. Old HP buttons feel much more solid and will probably last longer. Also much easier to repair or replace than chipped rubber keys.
« Last Edit: August 12, 2016, 05:25:54 pm by MadTux »
 

Offline garrettmTopic starter

  • Frequent Contributor
  • **
  • Posts: 279
  • Country: us
Re: Fluke 8506A Repair
« Reply #9 on: August 29, 2016, 11:27:02 pm »
I finally received the U20 replacement EEPROMs today.

After installation the meter throws "error d" (which, in this case, means no readable cal data stored on U20) but continues past and works just fine.

To correct this, set the cal switch to "on"** and then press  [Store]->[Zero VDC / Ohms]->[Zero VDC / Ohms] to clear the cal rom (that is U20), this may take a second or two.

Next, while still under cal mode, range down to 100mVDC* and short the input with some copper wire, wait for temp to stabilize (say 60 seconds or so) then press [Zero VDC / Ohms] and up range once and then press [Zero VDC / Ohms] twice (to set new zero correction) and repeat this process until you've reached the 1kV range, this restores the zero corrections for the DCV ranges (which are stored on U20).

Finally, set the cal switch to "off" and rest or hard power cycle the unit. You should no longer see "error d" and will now read zero volts DC on all ranges when appropriately shorted.

** You may see "error f" which is normal since there is no cal data stored on U20 yet. Proceed as indicated above and both "error d" and "error f" will go away.

* Optionally, adjust the zero and bias pots on the DC signal conditioner (as shown in the calibration section of the user manual) for a hardware zero on the 100mV range then proceed with software zero correction as indicated above.
« Last Edit: September 01, 2016, 06:16:06 am by garrettm »
 
The following users thanked this post: evb149, Storm, mdubinko

Offline ober

  • Contributor
  • Posts: 18
  • Country: de
Re: Fluke 8506A Repair
« Reply #10 on: January 28, 2019, 09:42:23 pm »
So does the line filter in a 8506A explode?  The service manual says this is a Fluke #649988, but I can't find any useful information about these or the capacitors in there.
 

Offline pelule

  • Frequent Contributor
  • **
  • Posts: 513
  • Country: de
  • What is business? It’s other people’s money
Re: Fluke 8506A Repair
« Reply #11 on: January 30, 2019, 07:28:22 pm »
I owned some years ago a fully equipped Fluke 8506A (with Amps, Ohm and GPIB board, unfortunately you can't use Amps and Ohms at the same time) for $150.
A nice but bulky tool (similar as the HP 3456A and the HP-3455A I am owning also).
Stability is excellent (even a bit less then the 3456A).
/PeLuLe