Author Topic: Fluke 87 low mA reading  (Read 9631 times)
Offline catcow1234Topic starter
« on: May 19, 2016, 07:00:18 pm »
I have a Fluke 87 (original model) which has served me well.  It's been well looked after and is only used on the bench.
Last week I noticed that the mA reading was incorrect.  For example, a 1.5v battery and 10R resistor gives the reading of only 120mA but if I put the meter in the 10A range than the reading is 150mA so the fault only seems to effect the mA setting.
I've double checked with a Fluke 79 III and that gives the correct current reading (150) in the mA setting so I know it's the 87 at fault.
I have checked the fuses which are fine, and cleaned the upper (metal) and lower (carbon) parts of the switch & changed the battery with no luck.
I do have the service manual but I can't see any way of calibrating the current range.
Can anybody offer any help / ideas?
Thanks
« Last Edit: May 19, 2016, 07:42:19 pm by catcow1234 »
Offline Cloud
  • Country: si
Re: Fluke 87 low mA reading
« Reply #1 on: May 19, 2016, 07:36:03 pm »
Maybe it is the resistance of the shunt
« Reply #2 on: May 19, 2016, 07:42:48 pm »
Wouldn't that effect the 10A reading too?
Offline retrolefty
« Reply #3 on: May 19, 2016, 07:56:02 pm »
Burden voltage value of the MA range?
Online MK14
« Reply #4 on: May 19, 2016, 07:56:39 pm »
I think there is a good chance your meter is working just fine.
On the milliamps setting, it has a resistance (shunt) of about 2 ohms. So 1.5V and 10+2=12 Ohms, can be about 120 milliamps.
The 10 Amp range has a very low resistance (shunt), so will give a current much closer to the 150mA.
Other meters can have different resistances (shunts), so can't necessarily be used as a comparison.
Really you should be using a constant (and ideally calibrated or known/accurate) current source to check it properly. This "rough" way with the battery is just confusing. The battery has an internal resistance as well, which varies.
Offline JFJ
« Reply #5 on: May 19, 2016, 08:15:11 pm »
... a 1.5v battery and 10R resistor gives the reading of only 120mA but if I put the meter in the 10A range than the reading is 150mA ...
On the 400 mA range, the burden voltage of a Fluke 87 is, typically, 1.6 mV/mA. On the 10A range, it's only 0.03 mV/mA. That  equates to shunt resistances of 1.6R and 0.03R, for those respective ranges. Adding 1.6R to your 10R resistor would reduce the current to about 129mA, but the effect of adding 0.03R would be far less significant.
Your Fluke 79 III must have a much smaller burden voltage.
Online MK14
« Reply #6 on: May 19, 2016, 08:24:29 pm »
This link seems to say it is 1.8mV/mA
https://physics.ucsd.edu/neurophysics/Manuals/Fluke/87______umeng0800
Then remembering to add the test probes resistance as well, it is probably about 2 ohms plus battery internal resistance.
« Reply #7 on: May 19, 2016, 08:25:56 pm »
Thanks for the replies everyone.
It looks like the burden voltage could be the issue.  I think I will try to get hold of Dave's uCurrent.
Thanks again  :-+
Offline JFJ
« Reply #8 on: May 19, 2016, 09:51:52 pm »
I found the 1.6 mA/mV in a Fluke 87 service manual on my laptop (perhaps it's for a different revision).
Offline ModemHead
  • No user-serviceable parts inside.
    • Mr. ModemHead
« Reply #9 on: May 19, 2016, 10:23:40 pm »
If you want to compare current ranges on two meters, just put them in series.  The current through both will be exactly the same.
Online MK14
« Reply #10 on: May 19, 2016, 10:26:39 pm »
I agree, it could be different revisions or something.
The following link, seems to be a fluke document, nicely discussing burden voltage.
http://www.fluke.com/fluke/uses/comunidad/fluke-news-plus/articlecategories/electrical/burdenvoltage
Offline retiredcaps
« Reply #11 on: May 19, 2016, 10:38:22 pm »
W2AEW's channel talks about burden voltage with the Fluke 87 and 79 III ...
Offline retiredcaps
« Reply #12 on: May 19, 2016, 10:45:29 pm »
The Fluke 79 III has 40mA and A input jacks.  If you are measuring more than 40mA, you should be using the A jack.
Offline retiredcaps
« Reply #13 on: May 19, 2016, 10:55:43 pm »
Martin's video on burden voltage ...
Offline VK5RC
« Reply #14 on: May 19, 2016, 11:28:37 pm »
You could make a simple constant current source with a LM317 and resistor,  that should remove the burden voltage effect.
Online MK14
« Reply #15 on: May 20, 2016, 12:15:37 am »
http://www.bristolwatch.com/ccs/LM317.htm
« Reply #16 on: May 20, 2016, 07:06:54 am »
@retiredcaps, Great Youtube link & you are correct, I was using the A input on the 79 III so the shunt would have been a different value to the 87 on mA so not a good comparison.
Thanks again everyone for all the help and advice.  I'll make the LM317 circuit later and give it a test.
« Reply #17 on: May 22, 2016, 12:56:09 pm »
I made the constant current circuit using the LM317 (thanks MK14) and the meter reads fine so it looks like I got caught out by burden voltage after all  :palm:
I compared it to a cheap Tenma meter and allowing for tolerances they both read fine (see attached pic).
Thanks again everbody for your help on this.
 :-DMM