I bought a cheap broken Fluke Series I (1986 Revision H) multimeter and am attempting to repair it. I found that the fusible 1k resistor was open so I replaced it with another 1k fusible resistor.  The meter appears to be working but the readings are off. I downloaded the Fluke 83, 85, 87 Series Multimeters Service Manual and followed the procedure for calibrating volts ac and volts dc. The problem is that I cannot get the readings low enough and the PCB pots don't appear to vary the reading much. For example when I put a 3.5 volt source into the meter I get a reading of about 5 volts.
Any advice would be greatly appreciated.
Title: Re: Fluke 87 Series I calibration issue
Post by: Axtman on May 18, 2019, 10:56:33 pm
The service manual is located here:
http://assets.fluke.com/manuals/83_85_87smeng0500.pdf (http://assets.fluke.com/manuals/83_85_87smeng0500.pdf)
Post by: retiredcaps on May 20, 2019, 02:55:43 am
An open fusible resistor suggests that more parts of the input protection could be bad.  Check the MOVs and PTC.
A good MOV should read "0L" in resistance.  A good PTC should read around 1K ohms.  Both measurements can be done in circuit.
Post by: retiredcaps on May 20, 2019, 02:56:42 am
No calibration will ever trim it from 5V down to 3.5V.
Post by: floobydust on May 20, 2019, 04:54:14 am
For voltage readings to be too high - that is worrisome.
If the 1k fusible resistor popped, the 1.5k PTC might be damaged, or the 909k resistor for mV and V. I'd guess it was overloaded on ohms or diode-test.
The diode/transistor clamps are switched-in for ohms and diode test functions.
On DCV the Fluke 87 is relying on the 909k input resistor - so leaky clamp diodes/MOV/transistors etc. I don't think can make it read higher?
I would check pin 2 AGND compared to input COM(-), the analog DMM IC has a few grounds. I would check the 1.235V reference near pins 1,3.
Post by: Axtman on May 21, 2019, 02:45:50 am
The 1.5k PTC (RT1) measures 1.24k.
The 909k input resistor (R2) measures 918k.
The two MOVs (RV1 and RV2) measure OL. Interesting enough one of the MOVs was not soldered to the board. It looks like it was a factory mistake. I soldered the MOV back in.
Post by: retiredcaps on May 21, 2019, 03:00:51 am
I forgot to mention to check the spark gap.  It should also read 0L in resistance mode.
Post by: Axtman on May 21, 2019, 03:05:38 am
Spark gap measures OL.
Post by: retiredcaps on May 21, 2019, 03:05:54 am
Put the Fluke 87 into DCV mode.
Using another meter, what is the input impedance of the 87?  It should be around 11.11 M ohm.
Post by: Axtman on May 21, 2019, 03:42:11 am
I get 10.28 M ohms.
Post by: retiredcaps on May 21, 2019, 03:56:12 am
It should be 11.11 M ohm. See
https://www.eevblog.com/forum/repair/fluke-87-i-repair/msg390462/#msg390462 (https://www.eevblog.com/forum/repair/fluke-87-i-repair/msg390462/#msg390462)
I suggest a complete IPA bath of the entire PCB.  Remove the lcd.  Let it dry completely and recheck the input impedance.
Post by: Axtman on May 21, 2019, 04:24:16 am
The pcb is very clean as is the rotary switch. There is no evidence of carbon, solder splatter, or other contaminants.
Post by: Axtman on May 21, 2019, 04:26:17 am
I checked the Pin 2 AGND with respect to COM (-) and got a reading of -1.728 v. Floobydust said it should be 1.235 v.
Post by: retiredcaps on May 21, 2019, 07:11:13 pm
What about all the other readings in table 2-1?  Please report them all.