Hello all, thank you for reading this. Possibly you can help me with your experience.
I have an old Fluke 8800A multimeter on my bench, with a somewhat complex fault in the DC area.
When in 200mV, 2V and 20V DC ranges, without anything connected, the offset slowly but steadily decreases, in the 20V range down to -4,5 V.
As soon as I short the input, it displays 0V as expected.
Also, switching to the 200V and 1200V DC ranges, without anything connected, it returns to 0V immediately.
Measuring inside, I saw that the voltages supplying U1 and U2 are largely offset, not symmetrical any longer.
When the 200V and 1200V ranges are selected, the relay K5 on the left connects the input to ground via 100kOhm, this is sufficient to bring everything back into balance. Even measuring voltage on the base of Q10/cathode of CR6 with another DVM (my good old HP466A) with its 10M input resistance is sufficient to have it back to working temporarily.
The 8800A has >1G Ohm input resistance in the lower ranges, this is probably what makes it so sensitive.
The troubleshooting section of the service manual mentions to shorten TP4 so that the bootstrap amplifier (U2/Q16) do not drive the DC buffer out of range. Doing that, the offset remains stable to some mV only.
I checked/replaced CR32/CR33 and Q16 so far, but no change.
Zeners CR11 and CR12 show a little deviation (one 5,9 V, the second 5,7 V). I don't know if this is important here. Have ordered replacements for them. My current thought is probably to replace U2, but that is an old OpAmp which is rare to get.
Possibly you have some experience with this kind of fault?
Thanks a lot from snowy Austria,
Hajo