Electronics => Repair => Topic started by: sycho123321 on November 15, 2015, 04:57:02 am

Title: Fluke 8800a repair
Post by: sycho123321 on November 15, 2015, 04:57:02 am
Hi everyone,
   Today I scored a fluke 8800a for 5$  :-+! It was known not working and I thought that it might be something really easy to diagnose and fix. And if non reparable, so what it would be a learning experience for 5 bucks. Straight away I downloaded the manual ( http://bama.edebris.com/manuals/fluke/8800a/ (http://bama.edebris.com/manuals/fluke/8800a/) ) And started working through the troubleshooting section. all the voltages were fine except for at section 4-52, f ,where it should be 5v +- 0.2v, but it was at 4v. I also went ahead and tested the adc (section 4-75) since on the schematic I did not see any 5v going into the adc. (I may be totally wrong!)  None of the tests passed. By the way I do have a o-scope I tested with. From what I can tell the adc is dead, right? unless the low 5v has something to do with it. On the display there are multiple decimal points no matter what range I am in, and there is always a mildly loud buzzing coming from the multimeter, sometimes it is louder on a few of the ranges.

Thank you very much!
Title: Re: Fluke 8800a repair
Post by: RJFreeman on November 15, 2015, 05:58:00 am
Quote
all the voltages were fine except for at section 4-52, f ,where it should be 5v +- 0.2v, but it was at 4v.

ok, the 5V rail is going to drive the logic in the multi-meter, and some older logic circuitry could be very sensitive to low supply voltages, and this could very well explain
Quote
On the display there are multiple decimal points no matter what range I am in,
so while it is only a volt low, i would try and sort this out first.
Also given it is making a buzzing sound, this could be indicative of excessive ripple, maybe due to dried out Electrolytics? what do these supply rails look like with your oscilloscope ?
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 15, 2015, 06:17:43 am
Thank you for the fast response! The rails look fine on my oscilloscope, and there are no telltale signs of dried out electrolitic caps. Also I forgot to mention in my origninal post that there is no visible damage on any of the boards. I think maybe the buzzing could be comming from  the main transformer.
Title: Re: Fluke 8800a repair
Post by: Shock on November 15, 2015, 06:43:58 am
I've not looked at the schematic but check everything on that rail, pulling current or connected to that component to be certain.
Title: Re: Fluke 8800a repair
Post by: tautech on November 15, 2015, 06:52:07 am
Not sure if you'll find something in this thread:
https://www.eevblog.com/forum/testgear/old-fluke-multimeters/ (https://www.eevblog.com/forum/testgear/old-fluke-multimeters/)

You could maybe ask the OP drtaylor, for advice.
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 15, 2015, 07:05:48 am
Thanks tautech for that awesome thread! And shock, thanks for the advice I will check around.
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 15, 2015, 07:15:20 am
Quick update, I removed the ohms converter board, and the main buzzing is gone but i've noticed a high pitched buzz which once I removed the ac converter board that buzzing stopped. But the problem of the ranging and reading of the voltage still exists. So i'm dong what shock suggested
Title: Re: Fluke 8800a repair
Post by: wiss on November 15, 2015, 10:33:15 am
I had a 8800 with a blown controller,  it did show those dots on the display if I remember correctly...
Fixed it by getting a controller from a 2100.
Title: Re: Fluke 8800a repair
Post by: Shock on November 15, 2015, 01:48:49 pm
Also low volts can be a problem with voltage regulation and filtering, again too lazy to check schematic.
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 15, 2015, 04:59:51 pm
I had a 8800 with a blown controller,  it did show those dots on the display if I remember correctly...
Fixed it by getting a controller from a 2100.

You're talking about u11 ,the main Ic, on the 8800a right

Edit: Oh the display controller, that would make sense, but all the ad/c tests have completely failed
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 15, 2015, 05:14:36 pm
Yeah I'm guessing either what shock said or  blown u11
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 15, 2015, 05:26:49 pm
Doh!! I don't think that what I was measuring is a 5v rail, the manual says to measure between tp9 and tp8 putting the ground lead of the multimeter to tp8 (which is -18v relative to ground) and the other lead to tp9 ( which is -14.3 relative to ground). So what I am effectivly doing is meesuring to difference. Why would the manual want me to do that?
Title: Re: Fluke 8800a repair
Post by: wiss on November 15, 2015, 05:38:31 pm
There is a bootstrap supply for the input buffer that can fail...
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 15, 2015, 06:22:16 pm
There is a bootstrap supply for the input buffer that can fail...

Just went through the testing procedure, It passed
Title: Re: Fluke 8800a repair
Post by: dacman on November 15, 2015, 08:52:42 pm
Doh!! I don't think that what I was measuring is a 5v rail, the manual says to measure between tp9 and tp8 putting the ground lead of the multimeter to tp8 (which is -18v relative to ground) and the other lead to tp9 ( which is -14.3 relative to ground). So what I am effectivly doing is meesuring to difference. Why would the manual want me to do that?

The 7805 regulator is referenced to the -18 Vdc rail.  The OUT and COM of the 7805 (U19) can be measured at TP8 and TP9.  Check the voltage across C22 (it should be at least 7.1 Vdc at the low end of the main line input voltage).  Check the voltage into CR31 (acV).
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 15, 2015, 09:05:17 pm
Doh!! I don't think that what I was measuring is a 5v rail, the manual says to measure between tp9 and tp8 putting the ground lead of the multimeter to tp8 (which is -18v relative to ground) and the other lead to tp9 ( which is -14.3 relative to ground). So what I am effectivly doing is meesuring to difference. Why would the manual want me to do that?

The 7805 regulator is referenced to the -18 Vdc rail.  The OUT and COM of the 7805 (U19) can be measured at TP8 and TP9.  Check the voltage across C22 (it should be at least 7.1 Vdc at the low end of the main line input voltage).  Check the voltage into CR31 (acV).

Thanks for the reply dacman, The voltage across c22 is a little low at 6.8vdc and the the voltage into cr31 is at 4.5vac. Should i pop a new 7805? as I have a few laying around
Title: Re: Fluke 8800a repair
Post by: dacman on November 15, 2015, 09:29:47 pm
The input to the rectifier seems low.  Can you check the position of S12 (115/230 Vac switch)?  I suppose the 5V output could be loaded.
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 15, 2015, 09:30:44 pm
the switch is set to 115v which is correct for where I live

I also swithed out the 7805 for good measure, still at 3.8 so looks like the input is low
Title: Re: Fluke 8800a repair
Post by: dacman on November 15, 2015, 09:35:25 pm
Those values seem low (from the transformer to the 5V output).  Try to exercise S12 a few times.  The 5V supply is either loaded or the transformer has low output.  Can the 5V supply be isolated?
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 15, 2015, 09:40:25 pm
I fliped s12 a few times, That didn't work. Also I think I may have inadvertantly measured the wrong in of the diode rectifier. But I have to go on an errand so I will be gone for a few hours. I will make sure I measured the right pin when I get back.
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 15, 2015, 11:41:25 pm
The input into the 7805 from the rectifier is 6.4vdc that seems low to me. And the input to the rectifier is 8.5vac So i did measure the wrong pins  :palm: . I think this means bad rectifier, or am I wrong?
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 15, 2015, 11:56:31 pm
Looks like something is loading it down, I powered  the 7805 i sucked out with 6.4v and wit is working fine.
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 16, 2015, 12:16:22 am
WE"VE GOT SOMETHING  :scared:  :-+ I revomved the 7805 and powerd the 5v rail from my bench power supply annd when I turned it on it showes overrange on everything, but the power draw max is only .37 amps, so nothing loading it down. But nomatter what I try none of the ranges will measure any thing but at least I temproarly fixed the 5v rail so I can trobleshoot this problem!!  ;D
Title: Re: Fluke 8800a repair
Post by: dacman on November 16, 2015, 02:09:23 am
I think your rectifier is bad.  The regulator requires at least 7 Vdc at its input (and because the input is unregulated, it should be 7 Vdc at 103.5 Vac line power input, else a line regulation check would fail, and at least 7.8 Vdc with 115 Vac line power input).  What is the Vac across C22?  The ripple usually goes up with failing rectifiers.  (The rectifier is a hard working part, continuously turning on and off with relatively high forward current.)  I'd also check those electrolytic capacitors.

Also, the input to these regulators normally have what may seem to be a lot of ripple, which the regulator must take out.  (They may work with a lower input voltage if the input voltage is smooth with relative little ripple, such as from a lab supply.)  Read the datasheet for your regulator.  Most passive regulators I've seen want 2 or 2.1 Vdc more voltage at the input than the output.  I've replaced several rectifiers due to low voltage at the regulator.  (And I want to know if the unit will work at its lower power line voltage specification, which can be directly tested using a Variac.) 
Title: Re: Fluke 8800a repair
Post by: dacman on November 16, 2015, 03:14:46 am
A generic cross for CR31 is KBP02 or 2KBP02 (I think).
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 16, 2015, 03:48:00 am
Thank you for those model numbers so I can replace the diode rectifier also the voltage across c11 is 11vac. Do you have any Idea what may be causing the 188888 blinking overload indacation on all the ranges? I have not had time yet to go through the rest trobleshooting in the manual yet.
Title: Re: Fluke 8800a repair
Post by: dacman on November 16, 2015, 03:54:19 am
11 Vac across (C22)?  Seems like C22 may be shot.  Do you have an LCR meter?  Please check all the electrolytic capacitors.  Troubleshooting with bad power supplies is futile.
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 16, 2015, 05:10:37 am
No, I don' but someone I know has one, will have to borrow it from them.
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 17, 2015, 04:29:33 am
I got ahold of a lcr meter but it only measures up to 2000mfd while c11 is 4000mfd
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 17, 2015, 04:59:39 am
Hmm when  I measure across c11 in acv it shows 1vac max  :-// the rest of the caps look good too. Maybe not a psu failure?
Title: Re: Fluke 8800a repair
Post by: dacman on November 17, 2015, 11:51:18 pm
What does it read if the input is shorted?  Are the four wire shorts attached?
Title: Re: Fluke 8800a repair
Post by: Shock on November 18, 2015, 01:03:01 am
Did you end up swapping the U19 regulator? Under no load and ideal input they may look fine but the moment they get a little load be symptomatic.

Diode test the rectifier diodes/pins in both directions. C22 has some change mentioned in the beginning of the manual. If it's smoothing to 7V then 10V seems a little small and likely why the supply has problems, I'd pick up a 16V as well as a 10V for comparison. Check what they used though it might have been changed to some other odd value already, you will have to scope it for ripple anyway to be 100% sure.

Check those grounds are common and then measure that against the rail voltages e.g. +/-18 and -13.
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 19, 2015, 01:27:23 am
What does it read if the input is shorted?  Are the four wire shorts attached?

Yes, there are  shorts on the banana jacks for the 4 wires, and it looks the same when shorted

Did you end up swapping the U19 regulator? Under no load and ideal input they may look fine but the moment they get a little load be symptomatic.

Diode test the rectifier diodes/pins in both directions. C22 has some change mentioned in the beginning of the manual. If it's smoothing to 7V then 10V seems a little small and likely why the supply has problems, I'd pick up a 16V as well as a 10V for comparison. Check what they used though it might have been changed to some other odd value already, you will have to scope it for ripple anyway to be 100% sure.

Check those grounds are common and then measure that against the rail voltages e.g. +/-18 and -13.

I did replace the u19 and the diode bridge is faulty, so right now I am powering the 5v rail w/ my bench supply. It is now showing overrange on all ranges so we know something is working. Also the grounds are not common. The ripple across c22 looks fine (I think the first time I measured it I did something wrong but it looks fine now)

Thanks guys!
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 19, 2015, 02:31:16 am
I just noticed also that the relay (k5) sometimes clicks sporadically when I change ranges.
Title: Re: Fluke 8800a repair
Post by: wiss on November 19, 2015, 07:24:33 am
Do you have the inputs shorted?
The relay should click when go from 20V to 200V and vv.
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 19, 2015, 11:42:29 pm
The bannana jacks have the shorting bars going to the sense inputs, But I do not have the his and ground inputs shorted, and the relay clicks sporatically even when not swithching between those ranges.
Title: Re: Fluke 8800a repair
Post by: wiss on November 20, 2015, 09:57:23 am
If it is in auto-range and inputs are open it will flip back and forth between 200V and 20V.
Title: Re: Fluke 8800a repair
Post by: halexa on November 20, 2015, 10:58:40 am
Hi,

I have an 8800A donor unit if you are in need of spare parts.

Good luck with your repair.
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 20, 2015, 10:45:49 pm
Hi,

I have an 8800A donor unit if you are in need of spare parts.

Good luck with your repair.

That would be AWESOME!
I will pm you,
thank you
Title: Re: Fluke 8800a repair
Post by: wiss on November 21, 2015, 02:01:14 pm
BTW, if it is autoranging, then the A/D probably works and the problem is upstream of that, i.e. something wrong with the display-circuit. 
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 21, 2015, 10:06:17 pm
It does not auto range I have been suspecting the adc, I think u11 is the adc right? Also when on auto range with the inputs open the relay does not click back and forth.
Title: Re: Fluke 8800a repair
Post by: wiss on November 23, 2015, 12:59:58 pm
U11 is the main controller, Q40, U4, U5 , U6 et.all are the integrator and comparator.
TP2 should have some kind of triangular waveform if the ADC is working.
Title: Re: Fluke 8800a repair
Post by: sycho123321 on November 24, 2015, 12:23:05 am
Yeah the ad/c is broken but I will ask halexa if he still has those components
Title: Re: Fluke 8800a repair
Post by: wiss on November 24, 2015, 11:09:34 am
U11 come in 2 different versions, there is a jumper on the PCB you will have to change if you install "the other one".
Title: Re: Fluke 8800a repair
Post by: halexa on November 24, 2015, 11:39:44 am
U11 come in 2 different versions, there is a jumper on the PCB you will have to change if you install "the other one".

How can you differ from the two versions?
Title: Re: Fluke 8800a repair
Post by: wiss on November 24, 2015, 12:29:13 pm
It is set by W1, there is a note in the schematic about it.
Title: Re: Fluke 8800a repair
Post by: halexa on November 24, 2015, 08:18:24 pm
Hi

Replacement parts U11, U4, U5, U6, Q40 and CR31 located and desolder ready for delivery to sycho123321.
Title: Re: Fluke 8800a repair
Post by: sycho123321 on December 05, 2015, 10:08:10 pm
Hey everyone!
I got the parts from helexa and soldered u4, u5, u6, q40, and cr31 in and cr31 did not fix the 5v rail so I am still powering that rail via a powersupply and it still showes overrange on all ranges, But since the u11 I got sent is the different version from mine I have to remove r121 and w1. Execpt that where the compnent overlay says w1 is there is a diode.
Any ideas?
Title: Re: Fluke 8800a repair
Post by: dacman on December 06, 2015, 01:33:28 am
Trace U11 pin 10 and RN4 pin 5.
Also, if you have a diode checker (function) what do the four diodes of CR31 test at in the forward direction?
Title: Re: Fluke 8800a repair
Post by: sycho123321 on December 06, 2015, 02:02:57 am
RN4 pin 5 goes to the diode i was talking about. U11 pin 10 goes to r121. I do not have a diode tester.
Lukas
Title: Re: Fluke 8800a repair
Post by: dacman on December 06, 2015, 03:55:39 am
Where does the other end of the diode go?  What is the AC voltage (center pins) into CR31?  What is the DC voltage at the outer pins?  What is the AC voltage at the outer pins?  For the diode checker, just use the 1 kOhm range of a DMM.
Title: Re: Fluke 8800a repair
Post by: sycho123321 on December 07, 2015, 01:03:21 am
the other end of the diode goes to pin 37 of u11