Author Topic: fluke 8808a  (Read 1015 times)
Offline duteletronicaTopic starter
  • Country: pt
fluke 8808a
« on: October 21, 2024, 10:05:06 pm »
Hello, I have a Fluke 8808a that was always measuring continuities with the beeping sound. After analyzing it, I came to the conclusion that it is the components E1 and E2 and I don't know what components they are? Can anyone help me find out the reference.
  • Country: pt
Re: fluke 8808a
« Reply #1 on: October 22, 2024, 02:18:48 pm »
please can someone help me???
Offline Swainster
  • Country: sg
Re: fluke 8808a
« Reply #2 on: October 23, 2024, 03:41:19 am »
Post a picture
  • Country: pt
Re: fluke 8808a
« Reply #3 on: October 23, 2024, 08:48:48 am »
Hi, here is the image. thanks
Offline picburner
Re: fluke 8808a
« Reply #4 on: October 23, 2024, 09:08:54 am »
They are part of the instrument's protection circuit, probably objects like these .
Offline picburner
Re: fluke 8808a
« Reply #5 on: October 23, 2024, 09:31:17 am »
.... and here  you can find some more clues.
In another user's instrument (although different in appearance) they are marked TAIYO AG20
  • Country: pt
Re: fluke 8808a
« Reply #6 on: October 23, 2024, 09:40:52 am »
Thank you, but is that the correct value??
Offline picburner
Re: fluke 8808a
« Reply #7 on: October 23, 2024, 10:12:18 am »
In the last link I posted the user GAD says this, I don't have this dmm so I can't confirm.
Offline Swainster
  • Country: sg
Re: fluke 8808a
« Reply #8 on: October 25, 2024, 03:59:42 am »
Just to be clear, spark gaps should read open circuit when doing a resistance or continuity test. You will probably need to pop out a leg to test them properly (just in case something else is in parallel with them). I don't think I've ever seen a failed spark gap without other obvious damage (burnt tracks etc).
Anyway, if you are sure that it is these parts that are faulty, then going by the red - black - red stripes,  BK12010002 looks like it might be a suitable replacement.
Offline picburner
Re: fluke 8808a
« Reply #9 on: October 25, 2024, 04:32:25 am »
One difference between SK20 and BK12010002 is that the first has a minimum insulation resistance of 10GOhm while the second has 100MOhm (if the datasheet of the latter is not wrong......).
For a dmm that on the 200mV/2V ranges has an input impedance > 10GOhm this can be an important difference.
The following users thanked this post: Swainster
Offline Swainster
  • Country: sg
Re: fluke 8808a
« Reply #10 on: October 25, 2024, 06:48:01 am »
Good point. I was just going by the appearance. Interesting that the plastic encapsulated part has better leakage characteristics than the glass one. Sadly I lost my old lab/workshop so my Keithley 237 has gone into storage, otherwise I could have tried to measure it. I think I have some old GDTs and SPGs left over from when I worked in the land line telephone business.
Offline picburner
Re: fluke 8808a
« Reply #11 on: October 25, 2024, 08:02:26 am »
Yes, I too would have expected a better device with glass insulation.
Looking more closely at the photo posted by the OP in the fourth post, only now I notice that the spark gaps do not seem to be factory mounted but would seem to be a subsequent repair, especially E2 gives me this impression.
  • Country: pt
Re: fluke 8808a
« Reply #12 on: January 03, 2025, 12:51:13 pm »
I thank you in advance for your help, it seems to have been repaired but the photo was only taken after unsoldering, I have difficulty in getting the component but I keep trying. Thank you