| Electronics > Repair |
| Fluke 8840A Faulty CPU |
| << < (4/22) > >> |
| essele:
Some slight progress... I’ve managed to find a disassembler for the Z8 (surprisingly hard to find a free one, finally found one as part of MAME, unidasm) and it’s really clear why the CPU code and prom code need to match ... the prom is just an extension of the CPU code, so it’s effectively a single 8K image. So even a minor difference won’t work as the location of subroutines will be different. My boards and components arrived (very impressed with JLCPCB again, and this was my first with KiCad!) ... although the programmer doesn’t appear to work! Just trying read I’m getting an overcurrent error if I use the TL866ii plus (set to a 13v Vpp) and if I use a GQ-4X then I get inconsistent results that aren’t all 0xFF (and it’s the same without the external 12.5v supply, so it doesn’t look like it’s latching into EPROM mode.) I need to double check all the programming voltages and then start looking at the timings, but re-reading the data sheet has worried me as there are some errors and inconsistencies in it. The other issue is that the Z86E40 doesn’t have a serial port (it has two comparators instead), this seems to be only used for the GPIB card, so that wouldn’t be a viable option with this solution, even if I could get it to program. |
| TheDefpom:
FYI Some time ago I received a faulty 8840A / 8842A unit and found it had the WRONG AC board in it, there is a connection difference between the two units, BUT the same AC board can indeed be used in both units, you just have to modify the interconnect wiring a little bit. I will see if I can dig that info up for you, I did note it down somewhere for my future reference, it might have even been in one of my repair videos. |
| TheDefpom:
Here you go, I found it. |
| essele:
Thanks Defpom - that is really interesting. That difference is very relevant since it also impacts one of the outputs of U208 which would be effectively shorted to GND when it shouldn't be, which could well be impacting CS7 and the associated watchdog. I have got three AC boards, one of which (the one from the AF) is fairly different looking to the other two so I will check the out. I'm assuming that pin 1 and 8 is also just swapped on the motherboard side (because of U208) so it's likely to be a pure hardware change and not a software related issue ... which doesn't explain why swapping CPU and PROM fixed it ... which probably means it's not my actual issue, but there are enough similarities that it definitely needs to be checked. Thanks. BTW ... I am losing the will to live with the Z86E40 ... will spend some more time this weekend, but I'm not hopeful about even being able to program it successfully. UPDATE: I checked all three and they are all the same, the boards match the first diagram (pin 1 connected to U803 pin 11) and the motherboards all have pin 8 directly connected to the CPU. So they would all seem to be compatible. I have now given up on the Z86E40, I have decided I hate Zilog, and I'm looking at the feasibility of a radical approach! More later. |
| essele:
Further info on my radical approach ... I'm going to try sticking in a PSoC4 running a kind of Z8 emulator! Initially I was considering a pure Z8 instruction emulator, however I don't think the performance will be there, the PSoC is 48MHz vs. the Z8 at 8MHz and the arm is pretty efficient, but for full emulation I don't think it's enough headroom to get close to native speed given the emulation complexity of some of the instructions that the Z8 handles in a very few clock cycles. [I'm not sure that performance is a particular issue in this case, so may be something to experiment with later.] So option 2 is to re-spin the firmware and effectively translate it into C with probably some assembly if needed, although the compiler will probably be better at optimising than I would be. I have written a simple disassembler in Perl ... stage 1 was to completely match the output of the MAME disassembler which is done (with the exception of the indexed LD instructions for which the datasheet is really vague and the MAME one seems to disagree with ... but thankfully it's not a problem.) Stage 2 was to update the disassember to actually follow the code paths so that I could work out which bits of the firmware were code and which were data (which is the reason why the indexed LD isn't a problem!) -- this is also done. The next step here is to update it again to isolate truly independent subroutines (i.e. ones that aren't jumped into by something else) so that they can be pulled out as separate functions, and then to start looking at detecting reads and writes to the control registers. The actual control registers look to be simple, RP is only set to 0xF0 in one place and is then followed by a load of RP relative code, the rest of the firmware uses a fully qualified read or write (for want of a better phrase.) So I'm working on having the disassembler identify RP relative access that can be fully qualified because no other code path gets in between the RP set and the access. The port access reads/writes are going to be a little more challenging and unfortunately some might need to be checked at run-time. There is a minor issue in the keyboard interrupt code since this uses a jump table in the firmware, so I think this is probably going to have to be manually re-written in C, although I want to try to avoid this if I can so the code translation is as automated as possible to cope easily with different versions. Other than that, from a software perspective, I think this is entirely doable. I would like to get my hands on other versions of the actual Z8 bit of the firmware. Hardware wise ... I've looked at the PSoC capability of driving the "external bus" using pure software, and whilst I think it might work, it's a bit fraught, it's at the limits of the performance of the device (using direct register writes, none of this API rubbish), interrupts would probably need to be disabled for most of the time, and it just feels wrong. I've put together a hardware bus component using verilog which is an approximation of the bus access process for both read and writes, mostly to see if it would fit within the available logic and it seems fine (with a fair amount of headroom), so it feels like that's the way to go. The good thing about the PSoC is that the same board could easily use the hardware or software option, and in fact reconfiguring the ports to not be address/data bus can be achieved through software so it would be possible to emulate that as well (although not needed for the 8840, other than a test mode I've found in the firmware.) The clock is an interesting point ... the IMO on the PSoC isn't as accurate at the crystal in the 8840, which may or may not be a problem. My plan was to use the IMO to generate the 8MHz clock internally and output it on XTAL2 so it can be used by the ADC (leaving XTAL1 unconnected.) I'm not convinced timing is critical (it's not an integrating ADC as far as I can tell) so this may be fine. There are three other alternatives ... (1) an external crystal on the board, (2) an external WCO to trim against, or (3) use the 8840's 8MHz crystal directly ... I think I can put together a board where all of these are an option, I just need to make sure it doesn't screw up the routing as some of the clock inputs are right in the middle of a contiguous port on the device I'm considering. Reset is another interesting consideration, if you use a normal pin for reset input you can choose to ignore the watchdog reset in the device which will simplify debugging. So next step on the hardware is to put together a board, I think I can squeeze a TQFP64 in between the DIP40. Once I have a board then I can start debugging in a live unit. Yes ... I know I'm mad. |
| Navigation |
| Message Index |
| Next page |
| Previous page |