Hi All,
I have a Fluke 8845A that is going out calibration. The following is what I was told by the person that calibrated it and it's history.
"Hey there Ed.
The Fluke 8845A sn: 9407014 has had a history of losing the calibration corrections made on the DC Volts function.
2008 our first year to calibrate the item, we received the unit being out (high) by 0.03% on all DC ranges.
We would go through the soft calibration and take our outgoing data and we can see the unit has come back into tolerance.
The next year we would receive it again and it would now be reading 0.04% high on DV functions only. Again we would go through the soft cal and get great results on our outgoing data.
Third cal cycle the DMM showed errors not only in DCV, but all other functions being 0.2 % high.
Then it was sent off to Fluke for Repair/Calibration, (Certificate # 2441853-9407014:1309112306)."
Has anyone had this problem before (or something similar)?
What components can be causing this? (Where should I look)?
Any help is appreciated. I also have the calibration data for this DMM. If that helps, I can attach them.
Thank you,
Ed
Seems the Firmware did not cure it long term.
When I took the data this morning, it nearly mirrored the 2008 data sheet for as far as I got.