Hmm...
Seems like this could be a common problem. :-BROKE
Jay
System error. Strike any user to continue.
Offline Johnny10
« Reply #26 on: October 06, 2018, 06:40:45 pm »
What does it cost to send them to Fluke for service?
Or is that a question not allowed in the repair forum. :-)
Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B
Offline GreyWoolfe
  • NW0LF
« Reply #27 on: October 06, 2018, 08:15:34 pm »
At least you didn't ask that question in the TEA thread!!
Offline DC1MC
« Reply #28 on: October 06, 2018, 08:20:39 pm »
The cost of sending them to Fluke for service is acceptable, on the other side, the cost of the servicing, this is what kills you ;).
Offline Jwalling
  • This is work?
« Reply #29 on: October 06, 2018, 08:43:01 pm »
 :-DD
Yep. IME, it costs about as much, if not more, than a used one in good working order. The four I have appear to have the same fault, but one is a bit different in it's behavior... More tomorrow.
Jay
Offline Johnny10
« Reply #30 on: October 06, 2018, 08:47:07 pm »
Yea, why did I even ask  :palm:
I should know better! Waaaaayyyyy too expensive.
I have two units with same error but one has all power rails working properly and one has shorted -15 and +15.
So same error covers a lot of different issues.
The following users thanked this post: analogRF
Offline giosif
« Reply #31 on: October 06, 2018, 09:01:17 pm »
Anyone getting bored of one of these sick puppies, I'd be glad to alleviate the burden.  ;D
Offline coromonadalix
« Reply #32 on: October 07, 2018, 03:01:10 am »
Sad to see theses meters having problems and Fluke does'nt acknowledge a thing ???
Offline TiNTopic starter
  • Country: ua
    • xDevs.com
« Reply #33 on: October 07, 2018, 03:33:01 am »
What would you want Fluke to do? :) I'm sure they will help you with repair for 90% of new unit cost  :-DMM
So far my experience with this 8846 just tend to consider it as overcomplicated design filled with unfriendly firmwares. BGA FPGAs with NIOS SoPC, no way to backup firmware, Fluke withdrawal of their own published FW and poor service documentation. Sure I'd accept that in meter like 8508A, but not in benchtop unit like 884X.
Offline Jwalling
  • This is work?
« Reply #34 on: October 07, 2018, 12:18:49 pm »
You're probably right. I can see the self test routine:
1 to 199: Run a whole bunch of self tests.
200: If anything fails, sit there and look stupid.
So the four units I have were bought by a friend of mine who brings me lots of repair work from a Govt. auction. Normally, he doesn't buy test gear, but other industrial equipment (hardness testers, etc)
https://www.govdeals.com/index.cfm?fa=Main.Item&itemid=17539&acctid=2863
If he had asked me about them before buying, I would have told him to stay away!
Symptoms:
1. All hang at the init screen.
2. When there is a USB thumb drive plugged in, three of them access the drive at 2 seconds and 8 seconds after powering on. No further accesses are done after that. The fourth one accesses the drive at 2 seconds, then at 3 second intervals over and over forever.
3. All voltage rails are verified and correct.
4. None respond to a GPIB scanning
5.All connect to LAN, and activity pulses can be seen at two second intervals. One time I saw one of them connect to my Asus router and get an IP address. I was unable to get it to repeat this though...
6. I never see any data going across the opto-coupler transmitter at DS3. The other one is transmitting a steady stream of pulses. I'll have to check whether the receiver is getting them, but you got to pull the board or remove the cover to access it. @ TiN: Does this sound like yours?
EDIT: The data stream from DS2 is making it across fine. Whether the data is correct, though... :-//
« Last Edit: October 07, 2018, 12:45:30 pm by Jwalling »
Jay
Offline Jwalling
  • This is work?
« Reply #35 on: October 07, 2018, 12:43:55 pm »
Shorted to GND? Do you have a 4 wire DMM?
Jay
Offline Johnny10
« Reply #36 on: October 07, 2018, 01:54:32 pm »
The 15 V rail sits at 2.18V the -15 at 0.434V
Found a comparator, IC LM393, that is running over 200 degrees F, on the second unit only 87degrees.
Will be investigating that today.
« Last Edit: October 08, 2018, 12:39:41 am by Johnny10 »
Offline Shock
« Reply #37 on: October 07, 2018, 03:03:26 pm »
Tin these look under done, please bake yours more. :D
Might be something to look at I cannot see from the thermal image what temp the right hand one is at. Did you check them and are those tantalums between?
Did you freeze spray that area at all? You may have hit a cap.
Offline Jwalling
  • This is work?
« Reply #38 on: October 07, 2018, 03:41:44 pm »
A better pic you can zoom in on:
https://xdevs.com/doc/Fluke/8846A/img/fixed_in.jpg
Jay
Offline HighVoltage
« Reply #39 on: October 07, 2018, 04:37:47 pm »
I used to have a 8846A because it was so much praised here on eevblog.
My unit was bought used with a very weak VFD, but I thought that would be an easy fix.
I tried forever to get a new VFD from Tektronix or other sources and I finally gave up and let go of the instrument.
  • Country: ua
    • xDevs.com
« Reply #40 on: October 07, 2018, 11:48:52 pm »
All power on my 8846 is fine, nothing cooking or shorted. It is not relevant to the "init screen hang" issue, because meter should boot even with inguard section completely unpowered (removed transformer cable from inguard).
Jwalling, yes, there is no life on optos from outguard to inguard. There is slow square wave on pair from inguard to outguard (perhaps just heartbeat signal).
I have tried so far things like:
* Replace RJ45 magnetics and Eth phy (suspected dead LAN input, that get NIOS CPU in FPGA stuck waiting) = no help.
* Replaced outguard SDRAM = no help
* Replaced outguard FPGA (bought new Cyclone from Digikey) = no help
* Freezespray bunch of stuff around  = no help anymore. Originally it helped and made meter work randomly for shortime.
When it was working, it took about 10-20 seconds to boot from init screen, and you can hear relay click after 2-3 sec from power on. Now it however does not click anything, just sitting silly.
Supervisor chip that drive reset to NIOS FPGA is ok, reset is correctly toggled high on power on. There is some life to SDRAM data pins, so it's doing something.
I tried to probe RS232 chip in hope to find debug CPU console output, but there was no signals.
So far best shot would be to trace JTAG to the debug connector, and connect USB Blaster, to try if anything can be done from that side (perhaps read back NOR flash).
But it's a long shot, as we don't have binary image from the FW flash. 8846A update tool have only bits and pieces, but not the full dump binary.
The following users thanked this post: Jwalling
Offline Jwalling
  • This is work?
« Reply #41 on: October 08, 2018, 10:40:15 am »
Re: Firmware for the Fluke version, have you tried contacting Erik?
https://www.eevblog.com/forum/testgear/fluke-8846a-anyone-that-got-the-updated-firmware-downloaded/msg1511872/#msg1511872
https://www.eevblog.com/forum/testgear/fluke-8846a-anyone-that-got-the-updated-firmware-downloaded/msg395578/#msg395578
Apparently, Fluke removed the update utility as too many people could not follow the instructions. Maybe they were bricking them?
assets.fluke.com/software/bench/readme.pdf
Jay
Offline Circlotron
« Reply #42 on: October 08, 2018, 10:46:41 am »
You could always couch it in terms of how much money you would be saving ;-)
  • Country: ua
    • xDevs.com
« Reply #43 on: October 08, 2018, 11:21:38 am »
Jwalling
You think where I got this one from?
Offline Jwalling
  • This is work?
« Reply #44 on: October 08, 2018, 11:39:24 am »
Jwalling
Oh. sorry. I thought you were looking for firmware...
Jay
Offline Jwalling
  • This is work?
« Reply #45 on: October 08, 2018, 11:51:36 am »
Doing some more poking around.
Cost to repair from Tek is $895.
The three units I have that behave the same with a USB drive attached, were from the University of Central Florida. All three had an RMA generated, but they never sent them in. Probably when they found out the repair cost.  :-DD
The fourth belonged to Entest Inc.
My gut feeling is that the firmware is probably corrupted on these. But it's just a guess...
Oh for the days when firmware was in PLCC sockets.
Jay
Offline Johnny10
« Reply #46 on: October 08, 2018, 01:06:19 pm »
$895.00 Guess that answers my question !
About the price of a used 8846A DMM.
But it comes back calibrated also!
« Last Edit: October 08, 2018, 01:34:04 pm by Johnny10 »
« Reply #47 on: October 08, 2018, 03:24:23 pm »
895$  Each dmm  or the loot ??
« Last Edit: October 08, 2018, 05:08:18 pm by coromonadalix »
Offline Kleinstein
« Reply #48 on: October 08, 2018, 04:16:23 pm »
I think from the screen shots it should be clear that this is $895 per unit and excluding shipping and likely sales tax.  Given the rather high price some units might end up as display donors. The repair price is so high, because is in many cases this would be a board swap.
Firmware corruption could be possible, though I would expect a check sum test at the beginning and than a corresponding error message - just like TiN's 3458. Several units with a similar fault might make it worth to dig a little further.  As a broken communication should be caught by the ground based part, I would suspect trouble with the outbound part. Just in case one could do a test with an intentionally broken link - at least this should be caught, as it would be an expected failure mode (e.g. spider at the optical link).
  • Country: ua
    • xDevs.com
« Reply #49 on: October 08, 2018, 10:33:10 pm »
Kleinstein
Usual NIOS people not thinking the same way as Motorola designers of 3458 FW. Why need to include debug/fault info for user, if we can just demo pretty Fluke intro bitmap on that lovely VFD, and ask user to pay for service? :)