This is from the (l)users manual 9.3 "User Maintenance"
You can revitalize a battery pack with too low capacity as
follows:
1. Reset the ScopeMeter test tool ( + ), and (ON-OFF)
ensure that the test tool is battery operated.
2. Press [MIN-MAX] to activate the Meter Min Max recording
mode.
This disables the battery save feature when no key entry is
made for 5 minutes. Wait until the test tool is automatically
turned off. When it is turned off, the batteries are
completely discharged. Depending on the condition of the
batteries, the discharging can last up to 8 hours.
3. When the test tool is turned off, connect the Power
Adapter/Battery Charger PM8907 to the test tool and
to local line power.
4. Charge for more than 21 hours.
5. Repeat steps 1 through 4 once more
The battery is not responsible for the program (it's in flash memory) memory only setup and time date RTC. I only have the 93, 95, 97 service manual but this is so differnt from the Fluke 91,92,96,99,105B ScopeMeter Serie II that I have so I think it's not to any help
Johan-Fredrik