Fluke 9100 Calibrator repair
« on: March 27, 2023, 04:53:52 am »
Our calibrator failed after unhappy connection to seems like an ammeter.
The following components were found burnt and replaced on the "9100 Analog PCB output switching, open sense detect & analog bus":
  • K012,
  • K003, K004,
  • F004
After that the instrument started issuing "SYSТЕМ TRIP: 9508 UNEXPECTED Power amp overload The Program Counter is 48EB0" on boot. And I don't know what to do next.

Have seen http://www.ko4bb.com/getsimple/index.php?id=manuals&dir=06_Misc_Test_Equipment/Datron/Datron_9100 but there aren't the description.

« Last Edit: March 28, 2023, 01:21:50 am by Machlee »
 

Offline Ugur

  • Regular Contributor
  • *
  • Posts: 84
  • Country: tr
Re: Fluke 9100 Calibrator repair
« Reply #1 on: March 28, 2023, 12:51:23 pm »
Your job is very difficult...
You can control the LM324 in the bottom corner on the right side of the power board.
 


Share me