Author Topic: Fluke PM 3394A self calibration  (Read 613 times)

0 Members and 1 Guest are viewing this topic.

Offline Hellfish69Topic starter

  • Newbie
  • Posts: 5
  • Country: at
Fluke PM 3394A self calibration
« on: December 26, 2021, 12:09:50 pm »
Hello,

i have scored a Fluke PM 3394A 200Mhz Scope. The Scope works fine. Self test ok analog mode works but in digital mode the scope has some issues.
Channel 1 and 3 have a big offset. The self cal runs but is ends with a incomplete massage. What can i do to fix this problem with the offset?
I removed the backup batteries (leaking) perhaps that causes the problem.


Lg Mike
 

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5479
  • Country: de
Re: Fluke PM 3394A self calibration
« Reply #1 on: December 27, 2021, 03:31:45 pm »
It has nothing to do with the backup batteries, they are only for RTC and memory of stored pictures and setups

Your problem is a hybrid chip that was originally manufactured by Philips and is now unobtanium.
The 4 Channel scopes have two of these chips, one for CH1,CH3 and the other one for CH2,CH4

There are 3 kinds of people in this world, those who can count and those who can not.
 