pizzigri:
Thank you Coromonadalix,

that is the draft I was talking about... I know that Fluke has an updated service manual for its version of the PM6666 (the one that officially supports 160 MHz instead of Philips' 120) that is over 120 pages long, and that is nowhere to be found.
That service manual is in fact branded Fluke....

schematics in the draft are good, however no silk screen, no part placement on the PCB and a pain overall the trace back parts on it....
FransW:
PM6666   operation manual (English) 1989.pdf, 2.8Mb,   70pages no schematics
PM6666  operation manual (English) 1996.pdf,   601kb,     46 pages, no schematics &
              "What Happens When I Switch to Local?", page = 33
PM6666  service manual (English) 1989.pdf,   6,8Mb, 41 pages with schematics & parts list

PM me & suggest where to send (if any)

Regards, Frans
coromonadalix:
i think the sv manual at 41 pages is the same as the one ive sent to ko4bb
FransW:
From the Fluke 1996 Manual:
FLUKE , Timer / Counter , PM 6666, 4822 872 20017 7/October 1996

GPIB-INTERFACE OPERATION.............................. …….23
Introduction .............................................................. 23
What can I do using the bus? ................................... 23
Connecting the controller.......................................... 24
Giving the counter an address................................... 24
Checking the communication.................................... 24
Two ways of programming ....................................... 25
Syntax........................................................................ 25
Selecting output separator........................................ 26
How to select function.............................................. 26

GPIB address-selector Location ..................................5
GPIB interface-connector Location .............................5
GPIB-INTERFACE OPERATION....................................23 - 36

What Happens When I Switch to Local?
Switching to LOCAL causes the counter to adapt the set- tings indicated on the display, see ‘How to select function’. This means that the counter will never have settings in LO- CAL which are not possible to set via the front panel.
When switching to remote again, the LOCAL-setting will re- main. Bus-functions like SRQ mask, output separator, EOI, etc. will not be altered by switching to LOCAL and back again.
Summary of Bus Commands Function Selecting Commands
FransW:
Manual (OCR) attached: