pizzigri:
Thinkfat,
I did as suggested! Yes, that is exactly what happens. I started the meter with the GPIB board connected and waited until the thing locked up. Then, i disconnected the SDA line, still locked up. Then the SCL, and it unlocked, all this obviously with the meter powered. I am  looking at the manual excerpts that FransW is posting, and the "Addr 10" is indeed the GPIB...
thinkfat:
What I can see is that the OCR didn't really increase the utility of the scanned document :-)
pizzigri:
Yes, haha!
In any case better than nothing at all...
In the meantime I connected my scope and got a 5mhz signal, quite jittery but that could be imperfect probe technique.
pizzigri:
Going to say something really stupid here but... could it be a ground related issue?
I see dispersion on the chassis, and the oooold Schaffner filter is still in place - the ones that exploded...
Suggestions?
thinkfat:
Ground is probably not a major issue, but the 5 MHz signal doesn't look too good. Does the documentation say anything about it? 2.54V peak is too low for a CMOS input (which I assume IC2 has). A CMOS logic high level starts at 3.5V. For TTL it would be OK. Is there anything specified about maximum rise time?

EDIT: I saw you use a 1X probe. That might get you into a bandwidth issue. Switch the probe to 10X and make sure it's properly compensated. Then redo the measurement.