Author Topic: Fluke Scopemeter 99 Series II no Picture  (Read 1341 times)
Offline mario_nksTopic starter
« on: March 19, 2021, 05:55:08 am »
Hello,
i have an Problem with the Fluke 99 Scopemeter.  Whe i switch i on, i can her the buzzer, an on autoset the relais clicking. With an optical interface i get the screen shot from the scope. I dont know how i could start. The Manual i found dosn't macht to my device.
I trid to do some messurment.
V1= 3,9V
V6 = 2,66V
V3 = 1,38V
V4 = -14,3V
V5 = -15,6V
V2 = -16,9V
I think this is okay?
When i chance the Contrast, the Voltages on the V1 - V6 are changing
Mario
« Last Edit: March 19, 2021, 07:51:02 am by mario_nks »
Offline EHT
Re: Fluke Scopemeter 99 Series II no Picture
« Reply #1 on: March 20, 2021, 09:02:39 pm »
Hi Mario, you should be able to find the service manual for this via Google. Certainly sounds like the display.
« Reply #2 on: March 29, 2021, 06:55:11 am »
I can't fin any service manual for the Scope!!
Have anybody the scematic for this untit? the Modell 105 must be the same.
I have definitive problems on my digital Board!
Mario
Offline chukin
« Reply #3 on: April 13, 2021, 02:16:31 pm »
Hi Mario
I am having the same problem as you - I have a dead Fluke 96B. I think 92B, 96B, 99B and 105B scopes have the same service manual but cannot find it online anywhere.
I asked for help elsewhere in the forum, so far no reply:
https://www.eevblog.com/forum/testgear/old-philips-fluke-97/msg3548276/#msg3548276
If I find a manual I will let you have a copy. Please do the same for me, thanks
Peter
Modesty - my only failing!