Hello Peter,
thanks for the feedback. I own the manual, but it is not much of help. The repair flow chart mentiones only defects on some discrete capacitors or issue with the rotary switch. All this is fine. There should be a clock-crystal, 32 kHz, but I still wasn't able to find that cystal on the board.
Welcome to the forum.
I don't know if my reply will be of much help. But here it goes.
The clock crystal is right above the battery under the part of the shield that is at the lower right side of the switch.
I repaired mine back in 2017 by re-flowing the front end chip, AP75, the one behind the battery on the backside of the board. The soldering did appear OK to me as well (except for one leg that I found un-soldered) and re-soldering fixed my meter. After that I re-flowed the microprocessor, upper chip, the one behind the display as well. It is working great since then.
And it is a good idea to follow the fault guide. Speaking of which it does say to follow the overall system check (paragraph 4-16) in case the display hangs up in power-up self test as you are experiencing. It also says that the possible components at fault could be R14, R8, R15, R16, VR1, C7, C8, Z1, CR2, U1.
Good luck with your repair. Its not a bad meter.