Recently I got a Fluke/Philips PM3285A in non working condition. Two of the mains noise filter capacitors were blown and the main power switch was broken (no blown fuses though), I replaced the capacitors and put another switch and the scope turned on and displayed signals.
However, the CRT display seems to flicker(?). Here's a video of the issue (input coupled to ground):
https://www.youtube.com/watch?v=aGJAJimcj2U&feature=youtu.beI think there might be an issue with the high voltage, but I haven't found similar problems by Googling, so I don't know how to further isolate the issue.
CRT controls seem to work fine. I can control focus, trace intensity, and readout intensity.
The following voltages can be measured on the motherboard, high voltage is controlled by the 13.9v rail:
Marked Value Measured Value
+13.9v +13.7v
-13.9v -13.7v
+5v (analog) +4.93v
+5v (digital) +5.17v
+25v +25.5v
-7v -6.5v
+120v +118v
-120v -118v
The CRT uses +14,3kV and -2,2kV, which I haven't measured yet, since I would have to borrow a HV probe to try and measure them.
I added the schematics that seemed relevant to the problem, but I have the complete service manual, so if other schematics are needed I can add them here.
Also, while changing the noise filtering capacitors I noticed that a capacitor (C6208 in the primary converter schematic attached) looked like it might be faulty. Since it couples to the power supply ground, could it be related to the CRT flicker problem?
Am I right in thinking this might be a high voltage problem?
I can't see anything obviously faulty, has anyone seen similar faults? If so, is this one tied to a more specific part of the circuit?