copied from the manual
VARIABLE SUPPLIES :
Since the operation of both variable supplies is identical, the following
discussion will deal only with the B variable supply.
The transformer secondary is rectified by the full-wave bridge
comprising D13 through D16’ and is filtered by capacitors C14 and
C15’. This serves as the unregulated input to U3’ an LM317 adjustable
three-terminal regulator. The output of U3 is set by fixed resistor R16
and variable resistor R17 the B ADJUST control. Capacitors C16’ C17’
and C18 accomplish additional filtering for the output, and improve the
transient load response. Diode D19 protects the regulator from the
current flow that would occur into its adjust pin if the output becomes
shorted, and diode D20 protects the regulator from the condition of its
output being raised above its input voltage.
Resistors R9 and R10 sense current flow through the supply, creating
a voltage drop that ultimetaly is used for the current reading on the
digital display. As there is a constant current flowing in the supply
circuitry even when there is no load attached, it is necessary to subtract
this so that the display shows true load current. This is done by the
constant current sink made up of transistor Q2' resistors R13' R14' and
R15' and diodes D17 and D18' Resistors R8' R11' and R12 scale this
resulting voltage to supply a 0.1mV / mA signal to display circuitry.
Similarly, resistors R18' R19' and R20 scale the final supply output
voltage to procedure a 1mV / V signal for the display circuitry.