Author Topic: Gould 1602 repair help  (Read 3620 times)
Offline noodlebearTopic starter
« on: February 11, 2013, 04:42:35 pm »
I've just acquired a Gould 1602 scope which has a problem. On power on, the power light comes on but there is no other activity. After leaving the scope running for a while and then turning off and on I occasionally see all the channel 1 LED's flashing and once saw the CRT flash.
The service manual is available at http://dl.dropbox.com/u/8513299/Gould%201602_1604_Service%20Manual.pdf
and schematics at http://dl.dropbox.com/u/8513299/Gould%201602%201604%20schematics.pdf
I have had a look at the power supply voltages and all seem within spec. VAC on multimeter shows minimal ripple however I don't have another scope to test this properly.
Going through the troubleshooting guide in the service manual the first thing to check is the 1MHz E clock on U1 (the main processor, MC68B09P). Since I don't have a scope to check this with I used the frequency counter on my multimeter. This shows a frequency of almost exactly 2MHz (not 1MHz).
I wonder whether this could be a discrepancy in the service manual since it seems exactly double? Does anyone have any ideas as to how I could check this more thoroughly?
Thanks
Offline scirturbo
Re: Gould 1602 repair help
« Reply #1 on: June 08, 2013, 01:35:05 pm »
I've repaired my 1604 several years ago and from what I can remember is that one logic circuit had a short and also several tantalium caps were low on resistance. It had also a out of focus image which was caused by dry caps in the powersupply. A couple of inputs had short/open components.  A good start will be replacing all the tant. caps.  Look for brown and hot caps.
Hope this helps........
Simon
Offline Skeluroth
« Reply #2 on: May 20, 2020, 11:48:29 pm »
Hello,
yeah, i know this is a old thread but i have the exact same problems as you're describing it and i could not find anyone else with this problem.
So i am wondering if you (or any other member) found a solution?