tonycox01,
Unregulated DC is always present at current limiters Q7 and Q4, when the 3325A is plugged in and the power switch is OFF. Same with the +5 Volt regulator input.
Q8 and Q11 are responsible for turning on the +/- 15 Volt regulators, when the power switch is in the ON position. Note the power switch pulls the turn-on line to ground when depressed. Proper power supply operation depends on that one grounding screw marked “NOTE”, being electrically bonded to the chassis.
Best to run a tap into that hole to clean out the threads and add a lock washer under the screw head.
The power supply turn-on line is held up by way of relay coil K1, sourced from the unregulated DC input to the +5 Volt regulator. This maintains the power supply in the OFF state. Monitor the voltage on this turn-on line during your load testing, to see if it suddenly rises from zero Volts.
You didn't mention anything about a test load on the +5 Volt regulator. Do note that crowbar SCR CR10 gets activated when the unregulated input exceeds approximately 17 Vdc, effectively shorting this unregulated supply. This unregulated DC would be another item to monitor during load testing.
If there are signs of corrosion present, indicating this 3325A has been stored in a damp location, then the integrity of electrical bonding between all castings may be suspect.
RF+ Tech