Hello all,
Work and life have delayed the continuation of the troubleshooting. As ledtester posted the manual, I found it independently. My unit has no letter after the 801 but the schematic is quite close, or almost identical. Components listed don't completely match but very close.
Today I concentrated on validating the power supplies, to ensure the basics are covered. From the 18-0-18 VAC supply rectified by BD202, after all the regulation it is quite close to the +15.0 / -15.0 VDC expected.
The supply for the 6-0VAC is converted to aprox +5.0 VDC at the emitter of Q204 which varies from +5.13 to +4.27 VDC depending on the scale selected.
R211, 212, 213, 214, 215 and 216 are exactly correct.
The supply 9-0-9 that eventually provides the +12 and +5VDC is pending. Hope to look into later today or tomorrow.
I am lost on some of the functionality of the LM723 variable regulator function. I remember using those to fix an output voltage with external resistors and if needed a power transistor to allow more current. Still trying to understand that part of the circuit.
I estimate the 3140 are the X10 and X100 amplifiers.
To some that suggested I measure the current at each scale with shorted leads, here it goes.
Scale Current (A)
2K 0.000 Meter might not be sensitive enough
200 0.001
20 0.01
2 0.1
0.2 0.000 Something wrong here
0.02 0.046
I never got the expected 1A in the lower ranges. I tested D401 outside the circuit and appears OK, BTW. 1.46VDC forward drop.
I will continue investigating, so please keep up the suggestions. Work and life sometimes don't allow a more fluid communication lately.
Thank you all.