I have recently treated myself to a Tektronix ADA400A differential probe. It is basically a TekProbe version of some older series 5000/7000 plugins.
I have started doing a performance verification, and found that while gain, zero offset and LF CMRR are excellent, the HF CMRR is way off. Like, "10x higher than it should be" off.
The manual [1] has a procedure for alignment, so I went through that. On page 47, HF CMRR is adjusted to be <1mV when applying a 100kHz, 1Vpp, 50Ohm terminated signal to both inputs. For my probe, it's more like 12-13mV (see attached screenshot) and I am unable to get it down further using CV4 and R18.
Sanity checks: I have verified +/- 15V and +5V on the board. It appears the last wire (Brown) does not carry -5V, but rather indicates the attenuation factor to the scope. For giggles, I have played around with different lead lengths on the inputs, but even a 2.5m length difference does not really change the picture other than introducing additional spikes from reflections. There is no visual damage to any component that I could see and no leakage from the capacitors.
EDIT: I am using the Siglent SAG1021I frequency generator for my experiments. While it does not have great rise times, it's the best I have on hand and it should not really matter at the frequencies we are interested in here, especially when the same signal is fed into both inputs.
What am I missing here?
[1]
https://download.tek.com/manual/070916402.pdf