Hi!
Both the X plates of the crt are symmetrically driven by identical cascode directly coupled DC amplifiers, so it shouldn't be too difficult to localise this sort of fault!
The simplest method is as follows:–
Switch the oscilloscope to it's SLOWEST timebase speed and connect your d.v.m. to the base of T2206, and switch the triggering onto auto so you get the spot free–running at it's slowest speed, turning down the intensity to make it just visible! As the spot runs from left to right, note the minimum and maximum voltage recorded on the dvm (if your dvm has min/max recording this is a great help) – if not, try and note down the figures as accurately as you can.
Once you've got a pair of voltage figures for the maximum and minimum base voltage on T2206, transfer your meter prod to T2207, this time rotating the "X Pos" control from one end to the other – obviously this is much easier because "X pos" is a manual control.
Once you've got a pair of minimum and maximum voltage figures for the bases of T2206 and T2207, subtract the lower from the higher figure, the difference is the voltage swing at each base needed to deflect the spot across the screen horizontally.
The most important thing that matters is that the voltage changes on T2206 and T2207 bases should be approximately the same, as you're deflecting one crt and it's D1 and D2 plates will have the same sensitivity!
If you measure approximately the same minimum–to–maximum voltage changes at both T2206 and T2207 bases by following the above procedure, then any trace distortion or non–linearity will be down to the final X amplifier stages (T2206, T2208 & T2210 for the D1 plate, and T2207, T2209 & T2211 for the D2 plate) and straightforward component checks on the resistors & transistors should bring any defects to light.
If the voltage swings at T2206 and T2207 bases are NOT approximately equal by using the procedure I suggested, then you need to make a further quick check of the output circuits, by lifting one end of D2204 and D2205, then using short pieces of wire to connect the anode end of D2204 to the anode point of D2205 on the printed circuit board and vice–versa – this changes the D1 and D2 plate amplifier connections to the X amp input and the X pos control over, and if both sets of amplifiers are
working correctly, the trace should show the fault you're seeing on the left hand side of the screen if the fault is before the X final amplifier.
It sounds a fairly complicated procedure, but it's the only quick way of differentiating between X amplfier and timebase failure!
Please try the above if you can and come back if you're still stuck!
Chris Williams