Author Topic: Hantek DSO5102P not measuring anything on both channels  (Read 823 times)

0 Members and 1 Guest are viewing this topic.

Offline D4no0Topic starter

  • Contributor
  • Posts: 19
  • Country: md
Hantek DSO5102P not measuring anything on both channels
« on: March 12, 2020, 04:59:14 pm »
Hello,

I just got an Hantek DSO5102P oscilloscope and out of the box I wanted to calibrate the probes. After I hooked the probe to reference signal it worked for 5 seconds and after that it stopped. The scope is not measuring any voltage at all, connected it to my power supply without any success, on both channels.

When I turn the self calibration it gets stuck at 2/38 step with the message: Self calibration failed, error: 0xc00. Any idea why it might happen?
 

Offline D4no0Topic starter

  • Contributor
  • Posts: 19
  • Country: md
Re: Hantek DSO5102P not measuring anything on both channels
« Reply #1 on: March 13, 2020, 06:28:58 pm »
Okay looks like I solved the problem. There was a loose wire from the power supply to the logic board.
 