Hey guys,
I swear I posted this last night and today it's gone? Moderators - did someone delete it? Or did I not actually post it?
Been dealing with "scope wars" lately. Had this old Heathkit IO-12 since college (I'm 45 now), it finally took a crap about 5 years ago so I bought a Tek 2235 off the site and shelved the Heath. 2235 took a crap last year and I spent a time recapping it and trying to bring it back to life to no avail. There's a three page thread on that pig here somewhere - it's going back on the site for parts, I'm done with switch mode power supplies in "more complicated than I need" scopes is the short story.
So, back to the IO. I figure it's all I need, and simple. I just want to look at waves to see where tube guitar amps clip, I don't need stupid bandwidth and I don't even need it to be calibrated or even close to it - I just use an RMS DVM across the amp output to measure wattage at clipping. When I shelved it, the issue it has was that the Y-axis would go spastic every minute or so and the screen would turn to noise - I'd flip the sync switch back and forth to get it to stabilize and then it would do it again in another minute.
So I just recapped it. The big multisection can and both of the 1600V guys on the bottom were leaking goo everywhere. I swapped all of the electrolytics and all of the film caps but I didn't check any resistor values yet. Fired it up and I get a trace (yay!), and a sine with a sig gen hooked up. The issue is the trace is only about an inch wide with the horizontal gain maxed. I have a BK747B so I tested V6 and V7 and they're fine. DeOxited all of the controls (they needed that badly!) and they all seem to do their job. Any ideas?