Hello all. I have been mulling over joining EEVblog for years... I suppose it's time I did.
For my first entry, I'd like to share a recent acquisition: an HP 214A pulse generator from 1963. I have a penchant (and a use) for test equipment and I enjoy collecting and restoring these items when possible.
This unit came to me for $30CAD and was in non-working condition. It's beautifully built (oh wait, it's HP from '63... there is no question about build quality).
I did exactly what you're -not- supposed to do and powered it up and looked for smoke (the previous owner had already done this recently so I figure if any damage was to be done, it had already happened). I indeed got a little smoke but it was, thankfully, moisture and dust burning off a glass power resistor (R341).. no harm done.
After determining the source of the excessive current in the resistor to be a shorted electrolytic (C318), I replaced that and carried on sorting it out. I could get action in all the stages except the output itself. The bias voltages were all over the place and it turned out the shorted cap destroyed all the Zener/Avalanche diodes (CR320,321,322,326) in the chain and also the hold-on transistor (Q301), and CR327, 328 in said transistor's emitter circuit. I was able to find reasonable substitutes in my collection except for the transistor itself which is specified as HP number 1901-0441... great...
Looking through an updated copy of the HP manual which indicates design updates and part changes, HP had changed to the 2N2190 and this shed a bit of light on things. It's a low-voltage high-speed (for the day) PNP Ge diffused-base Mesa transistor. Not something I have handy. The NTE substitute NTE160 for some reason has lower voltage, current, and power specifications than the 2190, so forget that (I hate using NTE anything (or ECG if you're an old fart) anyhow). So in went a 2N2907A! In the end this worked out just fine.
It took me hours to get my head around how the whole thing works and find its various troubles but I eventually worked it out... normally these things go fast but this one was interesting and different. On final calibration I was able to bring everything into spec but the 10V output range. C317 was acting up which is interesting as it is in parallel with the originally-shorted C318 but probably only a coincidence. The ceramic cap here was holding the bias to about 6V and letting it go no higher. Maybe a crack in the dielectric that had oxide making a diode or MOV-ish effect? I'm not sure. Either way, with that sorted it works now and I have a nice pulse generator to play with!
You can see in the scope photo I'm getting somewhere in the 10ns ballpark rise on 10V into 50ohms. A bit more on the lumpy side than I want. The big lump turns out to be a cable reflection (calculated nicely for the length and type of coax I was using) so maybe my terminator isn't so good or I had a bad connection, who knows.