Author Topic: Help needed on HP 54601A oscilloscope display problem.(SOLVED)  (Read 4730 times)
Offline JacquesBBBTopic starter
Dear All,
I have scavenged an HP 54601A scope which I will be trying to restore to full working order.
As you see, there is a display problem.
I will be looking to the power supply soon, but
Does any one have a  hint on where the vertical problem of the display comes from ?
Lettes  are also inverted.
Thanks !
« Last Edit: March 07, 2015, 05:00:41 pm by JacquesBBB »
Offline KJDS
    • my website holding page
Re: Help needed on HP 54601A oscilloscope display problem.
« Reply #1 on: March 07, 2015, 12:28:07 pm »
It's one of two caps on the display driver board. From memory one is 1000uF and the other is either a 22 or 16uF cap.
Either can cause the issue, it's easiest just to change both. From memory I don't think the schematic was available, but if you find it and post it up then I should be able to highlight the caps. If not then a photo of the board may jog my memory.
« Reply #2 on: March 07, 2015, 01:26:20 pm »
Thanks KJDS,
Meanwhile, I  found  also this post on the keysight site
where the problem is discussed.
http://www.keysight.com/owc_discussions/thread.jspa?threadID=17356&start=15&tstart=-1
All the schematics are also available on the site at
http://www.keysight.com/main/editorial.jspx?cc=FR&lc=fre&ckey=1517718&nid=-536900196.536898053.00&id=1517718
I attach here the Display schematics. I will try now to change these caps.
Offline KJDS
« Reply #3 on: March 07, 2015, 02:07:15 pm »
That's close to what I remember, but I suspect I did a different scope in the series. I suspect that changing all the electrolytics round IC 601 will cure the problem. If you've not got an ESR meter then that's the easy way forwards.
« Reply #4 on: March 07, 2015, 04:33:26 pm »
All the  voltages of the power supply were right inside the spec intervall.
A rubber pad had decomposed, and this is something I already saw on another
HP 54601A that I fixed not so long ago.
I look for the bad caps which were supposed to be C608 and C609.
I desoldered both. C609 (1000 uF) was OK : 876 uF , 0.68 ohm ESR.
While C608 (10uF , 16V) was damaged (not even readable on the ESR meter).
See the picture.
I inspected the other caps, but they looked OK, so I left them on place.
Of course, I could have changed everything,  but the other caps were 100V , and
I suspect that C608 was  too low voltage for its task. I replaced it with a 100V, 10 uF cap.
C608 and C609 are the two caps that are under the date on the attached picture.
After that, and a little bit of calibration, everything is working fine, like magic
(sorry for the fuzzy pic).
I spend more time then changing the fan, as the original one made a terrible sound,
but I had to adapt the new fan to fit in place.
Thanks again KJDS
« Last Edit: March 07, 2015, 04:35:18 pm by JacquesBBB »
« Reply #5 on: March 07, 2015, 04:36:49 pm »
And the global picture
Offline GreyWoolfe
  • NW0LF
« Reply #6 on: March 07, 2015, 04:59:15 pm »
Glad the repair turned out to be simple.  Now it is time to enjoy the scope.  Take care of it so you don't  :-BROKE
Offline R Lamparter
Re: Help needed on HP 54601A oscilloscope display problem.(SOLVED)
« Reply #7 on: April 23, 2017, 07:17:40 pm »
Another successful repair using this information.  Only the small 10 uf capacitor, C608 on the video board tested bad (abt 7 uf and v high ESR).  Replacing that fixed the problem.  Before and after photos attached.
Offline Bushougoma
« Reply #8 on: April 24, 2017, 12:09:27 pm »
This is a common issue on CRT TVs based on the symptoms and value of the cap I would say that was the vertical pump up capacitor.
It's often the repair for compressed vertical deflection. I try not to think about how many CRTs were disposed of because of a single bad capacitor.
« Last Edit: April 24, 2017, 12:13:24 pm by Bushougoma »