Author Topic: Help, what model is this component? Looks like a discharge tube  (Read 1101 times)
Offline daisizhouTopic starter
  • Country: cn
« on: February 23, 2023, 09:28:32 am »
Hello everyone。I came across a component that looks like a discharge tube,It looks like a diode, but I think it's more like a discharge tube
I didn't find its data, if you know the data or know the exact model please let me know, thanks
daisizhou#sina.com #=@
Offline jonpaul
    • IEEE Spectrum
Re: Help, what model is this component? Looks like a discharge tube
« Reply #1 on: February 23, 2023, 10:54:40 am »
Teccor Transients Absorbing Zener  TAZ
j
An Internet Dinosaur...
  • Country: cn
« Reply #2 on: February 23, 2023, 11:02:39 am »
Can you know the exact model?
Offline fzabkar
« Reply #3 on: February 23, 2023, 11:42:39 pm »
What is the application?
  • Country: cn
« Reply #4 on: February 24, 2023, 01:07:52 am »
They are at the input end of ECG lead line,These pipes sometimes conduct (100k-300k), but they are normal after heating, so I don't know whether the components are damaged or normal, so I want to know the exact model and query the parameters
Offline fzabkar
« Reply #5 on: February 24, 2023, 01:45:13 am »
As I understand it, the input protection is required to protect the instrument from the high voltages produced by a defibrillator.
https://www.medteq.net/article/iec-60601-1-defibrillator-protection-design-test
http://www.medteq.info/med/files/Image/defib/ECG.png
Apparently, the relevant standard is IEC 60601-2-27:2005.
Could these be bidirectional 860V devices (sidactor)? Or 86V devices?
Here is a complete catalogue from 2004:
https://web.archive.org/web/20040414135933/http://www.teccor.com/web/PDF%20Files/Telecom/Complete.pdf
https://web.archive.org/web/20021209163418/http://teccor.com/web/menuitems/downloads/datasheets.htm#ds_sidactor
« Last Edit: February 25, 2023, 12:39:27 am by fzabkar »
Offline jonpaul
    • IEEE Spectrum
« Reply #6 on: February 24, 2023, 03:47:27 am »
for sure Teccor trademark, the bidirectional Sidactor (TM) is a unique 4 layer protector ideal for,such use.
A TAZ or gas gap cannot be used.
In patient connected medical device like EKG, ECG, etc the liabilities, legal and compliance JP have,stick government requirements for the device design, manufacture and services.
Any part replacement, substitution or modification generally invalidites  the approvals.
Consult your country's medical device regulations
Jon
  • Country: cn
« Reply #7 on: February 24, 2023, 04:02:38 am »
Ok, I understand. Thank you for your help
  • Country: cn
« Reply #8 on: February 24, 2023, 04:05:35 am »
Ok,Thank you for your help
Offline fzabkar
« Reply #9 on: February 24, 2023, 04:26:51 am »
Manufacturers normally comply with the basic differential requirement by using a shunt arrangement: a component such as gas tube spark gap or MOV is placed in parallel with the leads, and shunts the energy away from the internal circuits.
I would have thought that a spark gap or MOV would be too slow. :-?