Hallo every one,When I was checking the PCB circuit board,I found a P4 potentiometer and a 150 ohm resistor missing from the schematic diagram. I don't know where the most suitable position is for it.If you know, please let me know. Thank you
This is the working principle of this part of the circuit
**ECG curves are generated by the microprocessor based on data tables. The processor updates the 8-bit D/A-converter in IC1 (channel A), usually 500 times per second. IC-3 converts current to voltage, and IC-2 makes the signal bipolar. The amplitude is adjusted using P1, and the zero point is adjusted using P4.
IC1 D/A-converter part B sets the reference level for the curve generator D/A-converter part A. Part B thus determines the amplitude of the stimulus signal.
The DC value for maximum amplitude is read back by the microprocessor's 10-bit A/D-input El, and can therefore be adjusted precisely.
The other half of IC2 is used for operating the high-level output or the ECG signal. The circuit has an amplification of 4 times.
The signal is filtered for high frequency noise with F18 before it is conducted to contact J8. C3 ensures bandwidth limitation in IC2. D9 is over voltage protection.**