tl;dr
Here my brief report of repair. My scope (DVS version) failed with excact the same symptoms. Distorted image, near all LED's lit, no reaction on any key or knob.
I've done the mistake in following the TEK fault-tree. At measuring port (empty IC socket) half of the Voltages were missing or too low (esp the negative ones).
The fault-tree always guided me to failures in PSU. So i ripped it out and recapped it. After an isolated burn-in on workbench it runs as expected. But putting it back nothing of instrument behavior changed.
Due to lack of any brilliant idea finding this fault, i decided to focus on abnormal power dissipation of components. In short, which of the hybrids are running hot?
Closest was the vertical output hybrid. After pulling it, most functions were back. So i replaced it with an used from bay.
Very disappointing! Same fault.
And my second mistake. Not consequently following the logic.
After that i pulled Z-axis hybrid. This gives one the chace to observe the function of Input-amps, Attenuators, horizontal and vertical stages (of course with an distorted image, Vertcal Lines, no blanking...). If i had swapped these steps, i had saved 50$.
So my recommendation is clear -pull the Z-axis hybrid and observe scopes behavior.
p.S. at moment the 3rd Z-Axis hybrid working in that scope
p.p.S. if you decide, repairing the PSU, be very careful at removing the fan wheel! It's a very delicate, fagile unit with no replacement. Here in eevblog exists an excellent howto thread.