Hello Guys ,
I'm new on this forum ...I have one problem with this osciloscop Fluke PM3380B , on Digital the trace locks ok but on Analog mode is strange  ???,
I attached 2 pictures (one digital mode and one Analog mode) maybe you have some ideea.
https://imagizer.imageshack.com/img923/2816/o79Ayv.jpg
https://imagizer.imageshack.com/img924/6990/5sR42k.jpg
Thank you very much!
Offline grumpydoc
Re: Repair FLUKE PM3380B
« Reply #1 on: March 06, 2021, 09:41:10 pm »
You are seeing the flyback.
Get the manual (search online, several copies) and go through the manual calibration procedure - which, amongst other stuff, sets the beam dark current. You'll probably find it's OK after that.
Offline TimFox
« Reply #2 on: March 06, 2021, 10:14:02 pm »
A more common name for that is "retrace", where "retrace blanking" may be found in the manual.  "Flyback" usually refers to a power-supply technology, although traditional TV sets generated the CRT high-voltage during the retrace period and that was called "flyback".  Just a suggestion to guide searching for the topic.
« Reply #3 on: March 07, 2021, 06:49:37 am »
Ok , thank you for your recomendation , I will chek this and I will come back with an answer :-+
Online HighVoltage
« Reply #4 on: March 07, 2021, 08:51:48 am »
yes, this is a setting in the calibration and maintenance procedure and a very simple correction, if everything else is alight.
I never had this problem isolated by itself.
Maybe someone played with the calibration settings?
To access the maintenance screen, you need to enter the password: "32415"
After you are happy with the setting, you need to confirm your changes with a push of a small tool though the pin-hole in the front panel.
(Behind the calibration seal)
Let us know, if you fixed it
« Reply #5 on: March 07, 2021, 10:20:24 am »
I made those calibration settings ,I pressed the pin-hole but it doesn't store anything, when I pressed the pin-hole on the screen I see 6 empty calibration ,soo the memory is ok
also when I start the osciloscop , on the screen i see two error , battery back-up missing and time is not valid :-//
Online HighVoltage
« Reply #6 on: March 07, 2021, 11:00:55 am »
Did you check the two AA batteries in the back?