HG 5502 scope repair
« on: January 10, 2015, 11:20:06 pm »
Hello everybody here.

Just found the forum on tapatalk and decided to join in.

Anyways , i have recently got an old scope from a friend .

I decided to check a pwm ic output with it and i accidentally shorted some pins with test probe .

Now i get this on the screen (see bellow)


i checked the probe resistance and its about 400 kohms.

could the problem be related to the probe ? Or worse?

 

Online tautech

  • Super Contributor
  • ***
  • Posts: 29614
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: HG 5502 scope repair
« Reply #1 on: January 10, 2015, 11:57:55 pm »
Welcome to the forum.
What waveform does the Cal output on the scope display?
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline pedroteckTopic starter

  • Contributor
  • Posts: 27
Re: HG 5502 scope repair
« Reply #2 on: January 11, 2015, 12:24:45 am »
Hi and thanks for the quick replay , i must be lucky here .

Is this the one on the picture ?

Also what is the normal probe resistance please ?

« Last Edit: January 11, 2015, 12:30:48 am by pedroteck »
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 29614
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: HG 5502 scope repair
« Reply #3 on: January 11, 2015, 08:47:43 am »
Yes.
This Cal output will produce a 1 KHz 1 V peak to peak square wave.
It is used to calibrate a 10x probe only.
The trimmer of the probe compensation is to be adjusted for a perfectly shaped square wave display.
So adjust the vertical attenuation for close to a full vertical display and timebase for ~2 div period (0.5 mS (500 uS)) and adjust the probe trimmer.
So a 200 mV/ div setting should display 5 div high for 1x probe and 20 mV/div setting will display 5 div for 10x probe.

Understand all this and it will indicate if your probes are OK.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline pedroteckTopic starter

  • Contributor
  • Posts: 27
Re: HG 5502 scope repair
« Reply #4 on: January 11, 2015, 05:53:41 pm »
Ok man.

I have assembled the scope , i will check the cal signal and be back with news.
 

Offline pedroteckTopic starter

  • Contributor
  • Posts: 27
Re: HG 5502 scope repair
« Reply #5 on: January 11, 2015, 09:35:14 pm »
Hi i have checked the 1v cal output and it actually showed up on the screen .

I dont how it happened because before desassembling the scope i could not display anything.