EEVblog Electronics Community Forum
Electronics => Repair => Topic started by: TimFox on July 14, 2019, 09:40:04 pm
-
While I wait for the manual that I ordered on eBay, can anyone tell me the function of each of the six calibration pots facing up on the top board? I have the version using an AD 545 op amp, not the older version with photo chopper.
-
I would say take their values before adjusting them, and maybe by trial and errors you could find their use ??
-
I now have the manual. In case someone finds this on search, here is the layout of the calibration pots on the top board, seen from above. Each sets the full-scale value. Hickok recommends adjusting at 80% of F.S., and using 400 Hz for AC voltage calibration.
REAR
R48 [50 mA DC]
R47
[0.05 V DC]
R37
[0.5 V AC]
R39
[1.5 V AC]
R41
[5 V AC]
R44
[50 V AC]
FRONT