EEVblog Electronics Community Forum

Electronics => Repair => Topic started by: TimFox on July 14, 2019, 09:40:04 pm

Title: Hickok 1605M voltmeter
Post by: TimFox on July 14, 2019, 09:40:04 pm
While I wait for the manual that I ordered on eBay, can anyone tell me the function of each of the six calibration pots facing up on the top board?  I have the version using an AD 545 op amp, not the older version with photo chopper.
Title: Re: Hickok 1605M voltmeter
Post by: coromonadalix on July 15, 2019, 01:19:39 am
I would say take their values  before adjusting them,   and maybe by trial and errors you could find their use ??
Title: Re: Hickok 1605M voltmeter
Post by: TimFox on July 23, 2019, 08:59:52 pm
I now have the manual.  In case someone finds this on search, here is the layout of the calibration pots on the top board, seen from above.  Each sets the full-scale value.  Hickok recommends adjusting at 80% of F.S., and using 400 Hz for AC voltage calibration.

         REAR

                    R48 [50 mA DC]

R47
[0.05 V DC]

                    R37
                    [0.5 V AC]

R39 
[1.5 V AC]

                    R41
                    [5 V AC]

R44 
[50 V AC]

            FRONT