hi guys!
i've bought 2 HP 54501A. old very-high-end scopes. 100Mhz but only 10Msa/s
one don't power on... and the other one says various error...
i'm going to post here about this repair.
scope that work bad:
various error, including a/d, memory and DAC error. smell like corrupted calibration data.
it's fault of the NVRAM (non volatile ram)... it has 2 litium battery... but not user or service serviceable...
so, i've found a compatible replacement DS1230Y-120+ from maxim. i've ordered 2 (maybe non-original) from china to be sure, but i'm going to use original and expensive ones from dallas semiconductor.
i've already removed the chip HP PC Troubleshooting... it's very difficult, for be sure to don't damage the board i've destroy the chip...
inside is all trough hole tecnology... zero smd
i've also clean the main board and the power supply.
next things to do
-replace the nvram
-clean all the case
-check the fan
-clean the HV CRT board
-disassembly the front panel and clean the soft button
-reassembly all
-calibrate it
-complete function check
for the scope that doesn't work at all:
there wasn't fuses. i've inser good fuse in it, powered on and.... smell burned electronic
but for 6-7 second it boot correctly and displayed same error of the other one scope...
so i've open it... disassembled the power supply board and... old electrolitic caps fluid everywhere!!!
so i've desoldered all this NO BRAND CHEAP CAPS (the power supply was OEM from a hong kong company...)
i've also clean the board in isopropilic alcool for 4 hour... now is clean...
i will buy new caps soon...
things to do:
-buy new caps
-solder new caps
-check the power supply voltage
-destroy and desolder the nvram
-replace the nvram
-clean all the case
-check the fan
-clean the HV CRT board
-disassembly the front panel and clean the soft button
-reassembly all
-calibrate it
-complete function check
--------
soooooooo
little funny "old-stile" things..
all though hole
the fan is "made in we-germany"!!!!!! i think dave will love this... it's before the fall of the berlin wall! great piece of history
the oscillator are from FOX
Faulty NVRAM (non-volatile RAM) causing the issues.
Found a compatible replacement DS1230Y-120+ from Maxim.
Removed the old chip, but accidentally damaged it.
Cleaned the main board and power supply.
Next steps: Replace the NVRAM, clean the case, check the fan, clean the HV CRT board, disassemble the front panel and clean the soft buttons, reassemble everything, calibrate the scope, and complete function check.
For the scope that doesn't work at all:
No fuses present, inserted new fuse, but smelled burned electronics and it shut down.
Booted correctly for a short time and displayed the same errors as the other scope.
Disassembled the power supply board and found leaking electrolytic capacitors.
Desoldered the old capacitors and cleaned the board with isopropyl alcohol.
Next steps: Purchase new capacitors, solder them in place, check power supply voltage, destroy and desolder the NVRAM, replace the NVRAM, clean the case, check the fan, clean the HV CRT board, disassemble the front panel and clean the soft buttons, reassemble everything, calibrate the scope, and complete function check.