The meters can be accurate, it's usually parallax error and gravity that makes them seem bad. Ensure the mechanical zero adjust is correctly set.
What is typical:
Trimpot in parallel to the meter movement (30k resistor) - cal the full-scale meter current i.e. 20uA
Trimpot in series with the meter movement (500R trimpot) - cal meter coil DC resistance i.e. 15,000 ohms
I would use a DC power supply and series resistor like 100k and go for full scale on the 0.03mA DC range.
That is closest to direct access to the meter's coil (it includes internal 30k shunt resistor). Use a DMM in series to measure the loop current. If the meter is out, then the coil 30k (33k in some pics) resistor needs to be adjusted or a clamp diode is leaky, assuming the 30k ohm shunt is accurate.
After that, you calibrate the meter's total DC coil resistance, not sure what it is for the AF-105 guessing 15k. That is the 500R trimpot setting the dividers' load resistance including the 7.8k and meter coil... or you can pick a DCV range and calibrate to that full scale- assuming that range's resistors are accurate. If it isn't, then other ranges would be worse for accuracy.