Author Topic: Hitachi v352 oscilloscope repair help  (Read 918 times)
Offline lame_turtleTopic starter
  • Country: gr
« on: August 24, 2020, 09:11:13 pm »
Hi
I got a v352 a few months ago that I wanted to restore. Channel two was blown, so I replaced the tr101 and everything worked again . I decided I would recalibrate the scope eventually so I recapped it and started using it.
 After a few days I hadn't turn it on, I fire it up just to see a sinewave on both channels without anything connected on the inputs.
I do have a ds1054z if a second oscilloscope is needed.
Any idea where to start looking ?
Thank you
Edit: I used the scope a bit more to understand what's happening.
All the controls/functions do work on the scope !
The only problem is the scope is only showing a sinewave. Not even a flat line when nothing is connected or the coupling set to gnd. This is happening on both channels.
Nothing is obviously burned up inside the scope.
« Last Edit: August 24, 2020, 11:02:12 pm by lame_turtle »
Offline shakalnokturn
Re: Hitachi v352 oscilloscope repair help
« Reply #1 on: August 25, 2020, 09:19:34 am »
A good starting point is to check all power supply voltages and ripple.
  • Country: gr
« Reply #2 on: August 25, 2020, 09:21:47 am »
Sorry I didn't mention. Psu checks out fine.  Tested it and checked for ripple.
Online tunk
  • Country: no
« Reply #3 on: August 25, 2020, 09:59:46 am »
What's the frequency and amplitude of the sine wave?
  • Country: gr
« Reply #4 on: August 25, 2020, 10:51:04 am »
Found it. Both k30a's where toast. I did have some spares and the scope is back giving a nice flat line when grounded.
I'll do a full repair report with photos in the next few days.  ;D
Thanks
