Well, I didn't think before that I could say "unit is beyound economical repair" about any of Keithley Instruments device I got.
But like they say, never say never. My kindness lured me into the trap, I thought I'd give some help to
Jxw6637 and get my hands on 6517A to get some content for my site.
I was not prepared at all for what was coming...


Unit was packed poorly, just janked into cardboard box with few foam sheets, on the side.

As one can guess from white getter spots, VFD glass is definately ruined. This photo was taken after some IPA cleaning already, it was filthy and dirty.
It's beyound me how some owners treat
instrument that cost more than $5000 USD.

What happen to mains jack?!


Removing covers for eye-watering sightseeing..


Digital boards look okay, no butchery detected there. Firmware is B13-version, RTC battery is good, no burnt parts, everything original.

Front panel is same as one used in Keithley 2002, with onboard royer DC-AC-DC converter to drive VFD glass elements.

You can see, there is no vacuum in the display. Did it happen from shipping damage due to poor packing, or was like this prior to that, I don't know.

Issue is clear, vacuum seal nipple is cracked and destroyed, letting the air in.

Nice AUGAT collet sockets for firmware chips and usual Motorola 68000 processor system.

This unit had blown transformer, so now it have replaced one. Transformer alone for K6517 cost over $600+ USD. Do not make mistakes like this plugging instrument switched to 110V into 230V mains.


No problems here on digital board.

I hooked my K2400 front panel board to digital board alone and powered it by Keysight E36312A. Board went online, taking +5V 0.7A. No problems with digital board.

As of analog board assembly...
I don't have words left to describe, so I'll just let you see on your own...