I have an HP 1660AS logic analyzer with power supply 'Computer Products XL130-3604E' where the +5V reference (test point in service manual) is ~+2.50V without load, with acquisition board as only load the +5V test point voltage drops to ~+1.14V. Adjusting the pot (+5.00V ADJ) next to the 5V test point have no effect.
The voltages on the PSU 20-pin output connector without load are as follows.
pin 1,2,3,4 +2.50V (expected +5.00V)
pin 9 +2.49V (expected +3.50V)
pin 11 -0.01V (expected -5.20V)
pin 13 +8.24V (expected +12V)
pin 15 -5.99V (expected -12V)
pin 17 +5.37V (expected +12V)
pin 18 -0.01V (expected -5.20V)
pin 19 +6.36V (expected +15V)
When the PSU started failing it was intermittent, it could work for like 20 seconds, all LA self tests passing, then fail again. Now it's "constantly failing" with output voltages as mentioned previously.
I have since then replaced all electrolytic capacitors (31pcs total), roughly 20% of the old ones were leaking, but the problem still remains. Also checked power transistors in-circuit which seem OK, at least no shorts between collector and emitter.
When measuring the diode bridge from the main transformer I get +2.50V, same as pin 1,2,3,4 on the PSU 20-pin output connector which is supposed to be +5.00V, so I'm assuming something must be wrong before the regulation, perhaps the diodes in the bridge or the transformer?
Here is a photo of the PSU linked...
http://modrobert.ddns.net/pics/HP_1660AS_PSU_XL130-3604E_top.jpg...and the small board soldered vertically in PSU:
http://modrobert.ddns.net/pics/HP_1660AS_PSU_XL130-3604E_side.jpgIf anyone have any ideas or what to look for it would be appreciated, power supplies are not my strong suit. I can't find any schematics for the PSU (XL130-3604E) when searching online, and the HP 1660AS service manual does not cover it.